Dear Rado,
To firmly answer your question, I tried 5 accessible ones of all 6 VPPs
at three different positions (#3, 30, 60). The behavior are the same,
with a variable degree. In deed, there was more than a month delay from
VPP delivery to installation due to engineer schedule. However, it was
sealed in the container. Thanks for the input.
Best Regards,
Jian
On 01/27/2017 01:29 AM, Danev, Radostin wrote:
> Hi Jian,
>
> If the CTF rings are moving outwards (make sure you are underfocus!)
> that means positive charging. The reason is most probably
> contamination on the phase plate film. If the heater is working
> properly (voltage ~ 4.0V, current ~ 25 mA) then the contamination is
> pre-existing, deposited during manufacturing, transport or handling of
> the phase plate. Are all VPP slots showing the same behavior?
>
> Cheers,
>
> Rado
>
> ------------------------------------------------------------------------
> *From:* 3dem [3dem-bounces at ncmir.ucsd.edu] on behalf of Sacha De Carlo
> [sacha.decarlo at dectris.com]
> *Sent:* Thursday, January 26, 2017 8:32 AM
> *To:* jamon
> *Cc:* 3dem at ncmir.ucsd.edu
> *Subject:* Re: [3dem] FEI-VPP phase change to opposite direction
>
> Could it be that the holder, with the new VPP, is no longer sitting in
> the back focal plane, where it should be?
> In any case, these are things to check with you FEI service engineer.
>
> All the best,
> Sacha
>
>
>
> On Thu, Jan 26, 2017 at 6:11 AM, jamon <jamonshi at gmail.com
> <mailto:jamonshi at gmail.com>> wrote:
>
> Dear Lister,
>
> We just changed our FEI Volta Phase Plate on Krios. But after
> activation, the phase changed to opposite direction, the FFT rings
> moves outward instead of inwards as it should be. The gold
> particles all looks all white as if in overfocus. I had never seen
> such behavior with previous two VPPs we had. Did anybody have
> similar experience? Thanks in advance.
>
> --
> Best Regards,
> Jian SHI
> Manager, Cryo Electron Microscopy Facility
> Center for BioImaging Science
> National University of Singapore
>
>
> _______________________________________________
> 3dem mailing list
> 3dem at ncmir.ucsd.edu <mailto:3dem at ncmir.ucsd.edu>
> https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem
> <https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem>
>
>
>
>
> --
> Sacha De Carlo, Ph.D.
> Business Development Manager EM
> Phone: +41 56 500 2190| Mobile: +41 79 354
> 4740|Email:sacha.decarlo at dectris.com <mailto:sacha.decarlo at dectris.com>
> DECTRIS Ltd. | Taefernweg 1| 5405 Baden-Daettwil| Switzerland|
> www.dectris.com <https://www.dectris.com>
>
> LinkedIn <https://www.linkedin.com/company/5067919>
> facebook <https://www.facebook.com/pages/Dectris-Ltd/623855944369304>
> <https://twitter.com/DECTRIS_News>
>
>
>
>
>
> Confidentiality Note: This message is intended only for the use of the
> named
> recipient(s) and may contain confidential and/or privileged
> information. If you
> are not the intended recipient, please contact the sender and delete
> the message.
> Any unauthorized use of the information contained in this message is
> prohibited.
>
>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://mail.ncmir.ucsd.edu/pipermail/3dem/attachments/20170127/e03dba30/attachment.html>
More information about the 3dem
mailing list