[3dem] Magnification anisotropy at low mag settings on Titan Krios

Max Haider haider at ceos-gmbh.de
Mon May 19 07:53:08 PDT 2014


Dear Marin,
I would like to correct your statement on the reasons of this observed 
distortion (a different magnification for two perpendicular 
orientations). As I could follow on the thread below this distortion 
depends on the magnification step and, therefore, it is definitely not 
caused within the Cs-corrector. When you change the magnification you 
only change the settings of the various lenses within the projector 
system and there is no change at all of any lens of the corrector. The 
point I made (or wanted to make) is that you need two stigmators at 
different locations in the column to modify this distortion by exciting 
one stigmator and to compensate it with a 2nd one at a different 
position. In order to compensate an existing distortion one has  to 
measure the power and orientation of the stigmators. There does not 
exist an automatic alignment tool because I guess there is a source of a 
2-fold astigmatism somewhere in the projector and there is no stigmator 
nearby. This astigmatism can be observed and you compensate it with, for 
example, an objective lens stigmator leading to a distortion. Which 
varies with the magnification because you change the excitation of the 
lenses in the projector and, hence, you vary the force of the obj. 
stigmator within this area. As Zhiheng wrote FEI people are aware but 
currently there is no short time solution.
There are stigmators in the Cs-corrector but I am afraid those are too 
weak to be used for a 7% distortion and, again, one would have to use 
them for each magnification step with a different excitation. This I 
stated briefly to be used to circumvent or to reduce the distortion but 
neither these stigmators nor the Cs-corrector in general are causing the 
observed distortion.
Best regards
Max.



Yu, Zhiheng wrote:
>
> Hi Marin,
>
>  
>
> Thanks for following up on this thread and sharing the very useful 
> feedback. The 2% or so magnification anisotropy/image distortion Niko 
> referred to was from a Titan Krios without a Cs corrector. The main 
> suspect is the projection system and FEI told us that they are still 
> working on a solution.
>
>  
>
> Best Regards,
>
> Zhiheng Yu
>
> Manager of CryoEM Shared Facility, HHMI Janelia Farm Research Campus
>
>  
>
> *From:* 3dem-bounces at ncmir.ucsd.edu 
> [mailto:3dem-bounces at ncmir.ucsd.edu] *On Behalf Of *Marin van Heel
> *Sent:* Monday, May 19, 2014 8:54 AM
> *Cc:* 3dem; Max Haider
> *Subject:* Re: [3dem] Magnification anisotropy at low mag settings on 
> Titan Krios
>
>  
>
>
> Dear All,
>
> I have spoken to Max Haider about the issue and he told me there is a 
> straight-forward reason for this magnification behavior with the Cs 
> corrector.  The combination of a quadrupole lens at the beginning of 
> the corrector and another one at the end of the corrector can jointly 
> cause this effect.  On our NeCEN Cs corrected instrument, we measured 
> up to 7% magnification difference in two orthogonal directions in the 
> first data sets we collected with a new Falcon-2 camera on our 
> Cs-corrected instrument. This large difference in magnification has 
> since been corrected by re-adjusting the corrector but this issue 
> remains a concern for high resolution work. In particular, Rishi and 
> Sacha found that the fine tuning of the Cs Corrector is only 
> effective/stable after the high tension has been on and constant for 
> some 10-12 hours.
>
> Cheers,
>
> Marin
>
> On 04/04/2014 08:10, Marin van Heel wrote:
>
>
>     Dear All,
>
>     This could also be due to astigmatism in the illumination system.
>     Such astigmatism would mean that the illumination is not parallel
>     to the optical axis in (at least) one direction leading to
>     anisotropic magnification effects as per our paper:
>
>     G. van Duinen,  M. van Heel, and A. Patwardhan,* Magnification
>     variations due to illumination curvature and object defocus in
>     transmission electron microscopy, */Opt. Express /*13 *(2005)
>     9085-9093*. *
>     Hope this helps,
>
>     Marin
>
>     =======================================================
>
>     On 04/04/2014 02:17, Nikolaus Grigorieff wrote:
>
>         Dear Colleagues,
>
>          
>
>         We have recently noticed a problem with anisotropic magnification on one
>
>         of our Titan Krios microscopes. When recording data at a nominal
>
>         magnification of 29,000x, there seems to be an image distortion that
>
>         produces variable magnification in different directions of the image.
>
>         These variations were estimated using diffraction from gold particles to
>
>         be about 2%, a significant amount especially when working on large
>
>         assemblies such as viruses. The distortions can be approximately
>
>         corrected using image interpolation but this is not desirable, of
>
>         course. In one case, the resolution of a 700 Angstrom virus
>
>         reconstruction with data collected on a Gatan K2 direct electron
>
>         detector improved from 7 to 4 Angstrom after correcting for the distortions.
>
>          
>
>         The severity of the distortion depends on the magnification setting. At
>
>         37000x magnification the magnification anisotropy is about 1% and
>
>         59,000x it appears to be undetectable. Since most Krios microscopes are
>
>         only calibrated for magnifications of 59,000x and higher, it is possible
>
>         that the problem we have observed also occurs on other instruments. This
>
>         will be particularly relevant for instruments that operate with the K2
>
>         detector mentioned above. The pixel size of this detector (5 microns)
>
>         usually demands magnifications settings of 29,000x and lower where the
>
>         distortions are significant. Users of detectors with a larger pixel size
>
>         (e.g. the Falcon direct electron detector) are less likely to experience
>
>         the distortions since they will typically use magnifications of 59,000x
>
>         and higher.
>
>          
>
>         FEI have acknowledged the problem but at this point the cause is not
>
>         clear. We hope that the distortions can be corrected with a simple
>
>         recalibration of the projector lenses. We would be grateful if other
>
>         Titan Krios users could share their experience and possibly check if
>
>         distortions are detectable at lower magnification settings.
>
>          
>
>              Thanks,
>
>          
>
>                   Niko.
>
>         _______________________________________________
>
>         3dem mailing list
>
>         3dem at ncmir.ucsd.edu <mailto:3dem at ncmir.ucsd.edu>
>
>         https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem
>
>      
>
>
>
>
> -- 
> ================================================================
>  
>     Prof Dr Ir Marin van Heel
>  
>     Professor of Cryo-EM Data Processing
>  
>     Leiden University
>     NeCEN Building Room 05.27
>     Einsteinweg 55
>     2333 CC Leiden
>     The Netherlands
>      
>     Tel. NL: +31(0)715271424 // Mobile NL: +31(0)652736618
>     Skype:    Marin.van.Heel
>     email:  marin.vanheel(A_T)gmail.com
>     and:    mvh.office(A_T)gmail.com  
>  
> ----------------------------------------------
>  
>     Emeritus Professor of Structural Biology
>  
>     Imperial College London
>     Faculty of Natural Sciences
>     Biochemistry Building (Room 512)
>     South Kensington Campus 
>     London SW7 2AZ,  UK
>     email:  m.vanheel(A_T)ic.ac.uk 
>  
>     Tel. UK:   +44(0)2075945316 //Mobile: +44(0)7941540625
>  
> ----------------------------------------------
>     Visiting Professor at:
>  
>     Laborat�rio Nacional de Nanotecnologia - LNNano
>     CNPEM/ABTLuS, Campinas, Brazil
>     Brazilian mobile phone  +55-19-99369051  
>  
> ------------------------------------------------------------------
>  
> I receive many emails per day and, although I try, 
> there is no guarantee that I will actually read each incoming email. 
> Moreover, our Spam filters can be strikt and sometimes make 
> legitimate emails disappear (try the gmail accounts, alternatively)
>
> -->
>

-- 
=======================================================================
Dr. Max. Haider
CEOS GmbH,    Englerstr. 28, D-69126 Heidelberg