[3dem] [EXTERNAL] Cooling water

Thejusvi Ganesh tganesh at nimgs.com
Fri Feb 9 15:14:35 PST 2024


Hi Ruben! Have you had the water tested for pH? What is your policy for
flushing the lines and chillers? Aside from the usual filter cartridge
replacement and cleaning of the mesh filters, we will typically try to
drain/flush/fill the chiller loops during the annual TFS PM using bottled
distilled water. We used to have to backflush the K2 camera to dislodged
some buildup but have not had any issues with the K3s.

On Fri, Feb 9, 2024, 2:32 PM Ruben Diaz <pindusito at gmail.com> wrote:

> Hi all,
>
> I have a silly question: what type of water do people use to keep the
> microscopes happy and chilled? We have already had two instances of
> corrosion on the cameras (K3), which have resulted in uber-pricey repairs,
> but the other parts of the microscopes seem perfectly fine and
> corrosion-free. The water does not show any unusual hardness or presence of
> ions that could be blamed for the corrosion. I am very interested to hear
> if other people have faced similar problems, and how they have been solved.
>
> Ruben.
>
>