[3dem] BSL2 vs BSL1

Sheemei Lok sheemei.lok at duke-nus.edu.sg
Sun Jul 5 23:27:16 PDT 2020


Thanks Matthjin, how about if there’s a difference in BSL classification between different countries? For example, the rest of the world that particular pathogen is classified as BSL2 agent while in Netherlands, BSL3. Should there be an increased in maintenance fees?

Regards,

Sheemei

Shee-Mei Lok
Professor/National Research Foundation Investigator
Program in Emerging Infectious Diseases
Duke-NUS Medical School
KTP Building
Level 9
8 college Road
Singapore 169857
Ph:65-65165840

Professor/Principal Investigator
Center for BioImaging Sciences
Department of Biological Sciences
National University of Singapore
Singapore



From: 3dem <3dem-bounces at ncmir.ucsd.edu> on behalf of Matthijn VOS <matthijn.vos at pasteur.fr>
Date: Monday, 6 July 2020 at 1:53 PM
To: jamon <jamonshi at gmail.com>
Cc: "3dem at ncmir.ucsd.edu" <3dem at ncmir.ucsd.edu>
Subject: Re: [3dem] BSL2 vs BSL1



- External Email -



Maybe this helps

https://urldefense.com/v3/__https://consteril.com/biosafety-levels-difference/__;!!Mih3wA!SEBSw2muOb8jMoAkiyu66LUTbu7HxuNygfn7_oVtGYsRzL3bBcevo5xHl9z94ZsZrQ$ <https://urldefense.com/v3/__https:/consteril.com/biosafety-levels-difference/__;!!Mih3wA!XBbjyfi0Xo0lQGQ95joxA2DQP_Vfnq58ibMOr6C_s9FMgYf9OkoPYLvt9vuqsn35Fg$>

The difference in lab setup between BSL 1 and 2 are not that big. A lab manager can also decide to declare a lab BSL2 for a period of time, then decontaminate and resume operation as BSL1. TFS only charges extra service cost if microscope decontamination procedures are involved and parts can not be taken back. In BSL3 all parts stay in the facility until the lab closes. This means a company can not recover parts and reuse them. This will increase cost. Also decontamination procedures might damage certain parts that need to be replaced (other than the broken parts that needed fixing), which again increases cost. For BSL2 this does all not apply. Wearing lab coats, gloves, presence of a sink, ways to dispose of infectious material, a safety cabinet and modifications to the air flow I believe are required but are all “reversible” back to BSL1. There are no specifications based on disinfecting the inside of the microscope. Best is to contact your local TFS agent, they have a full unit for dealing with this, there are many BSL2 facilities worldwide that house EM.

Cheers


On Jul 6, 2020, at 04:21, jamon <jamonshi at gmail.com> wrote:

I heard there will be ~10% increase of service contract from BSL1 to BSL2. Has it been applied world wide?

Regards,

Jian
On 6/7/20 9:21 am, Sherman, Michael B. wrote:
Hi Ruben,
Jaap is correct, BSL2 assumes standard lab practice, lab coat, gloves at the most, but no special safety/decon procedures. And no service cost increase. TFS does not want to deal with BSL3, so that is out of question. Our Krios is classified by TFS as level 2 instrument and my engineers come and go as they like (face mask is required currently… for a different reason). I do not see any cosequences declaring your microscope s BSL2.
All the best,
Misha

Michael Sherman, PhD
Dept. of Biochem. & Mol. Biology
1.224 Medical Research Building
Univ. of Texas Medical Branch
Galveston, TX 77555-1055
Tel. 409-772-6310
Fax: 409-772-6299
mbsherma at UTMB.edu<mailto:mbsherma at UTMB.edu>


From: 3dem <3dem-bounces at ncmir.ucsd.edu><mailto:3dem-bounces at ncmir.ucsd.edu> on behalf of Jaap Brink <jbrink at jeol.com><mailto:jbrink at jeol.com>
Date: Sunday, July 5, 2020 at 8:10 PM
To: Ruben Diaz <pindusito at gmail.com><mailto:pindusito at gmail.com>
Cc: "3dem at ncmir.ucsd.edu"<mailto:3dem at ncmir.ucsd.edu> <3dem at ncmir.ucsd.edu><mailto:3dem at ncmir.ucsd.edu>
Subject: Re: [3dem] BSL2 vs BSL1

WARNING: This email originated from outside of UTMB's email system. Do not click links or open attachments unless you recognize the sender and know the content is safe.

Forgot to add that before SVC enters the facility, a wipe down is required. Akin to what you’d do with a lab bench.
Regards,
Jaap
----
Jacob Brink Ph.D.
TEM Product Manager Life Sciences
JEOL USA, Inc.,
Peabody MA 01960, U.S.A.
Cell: +1 (978) 317-9561

Sent from a mobile device. Please excuse typos and this email’s brevity.



On Jul 5, 2020, at 8:07 PM, Jaap Brink <jbrink at jeol.com><mailto:jbrink at jeol.com> wrote:
Hi Ruben,

Don’t think BSL2 requires that. BSL3 for sure does and requires elaborate decon procedures to be in place. UTMB goes for a SVC visit back from level 3 to level 2. It is then safe to enter.
Regards,
Jaap
----
Jacob Brink Ph.D.
TEM Product Manager Life Sciences
JEOL USA, Inc.,
Peabody MA 01960, U.S.A.
Cell: +1 (978) 317-9561

Sent from a mobile device. Please excuse typos and this email’s brevity.



On Jul 5, 2020, at 7:31 PM, Ruben Diaz <pindusito at gmail.com><mailto:pindusito at gmail.com> wrote:
Hi Jian,

Yes, if we decide to put BSL2 specimens in the microscopes, we have to tell TFS about it, to make sure that their engineers are safe.

Ruben.

On Sun, Jul 5, 2020 at 5:28 PM jamon <jamonshi at gmail.com<mailto:jamonshi at gmail.com>> wrote:

Dear Ruben,

Is it request by TFS (FEI)?

Regards,

Jian
On 6/7/20 7:17 am, Ruben Diaz wrote:
Hi all,

Due to the type of samples that we are interested in imaging, it would be advantageous for us to declare our facility as BSL2. This designation, however, has some implications in terms of the paperwork and safety measures that have to be taken whenever service is called. Cost of course I am sure is another factor. Does anyone have experience as to the real implications that have to be considered? It is my understanding that once we declare the facility to be BSL2, every instrument is included in the designation, and it is not reversible.

Any information will be greatly appreciated.

Ruben.

***************************
Ruben Diaz Avalos,
La Jolla Institute for Immunology,
9420 Athena Cir, La Jolla, CA 92037<https://urldefense.com/v3/__https:/nam11.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Fwww.google.com*2Fmaps*2Fsearch*2F9420*Athena*Cir*2C*La*Jolla*2C*CA*92037*3Fentry*3Dgmail*26source*3Dg__*3BKysrKysr!!Mih3wA!VF5pTkhY_jgbypbTZ84Z_wII5hzOEDg6Q_xEEk1iMjZ0w-xXRhuHwYXCownxYUuv-A*24&data=02*7C01*7Cmbsherma*40utmb.edu*7C7588ba702e3d48b0331608d82149701a*7C7bef256d85db4526a72d31aea2546852*7C0*7C0*7C637295946590205739&sdata=gf0ueEnZLbDRuXn9RIx8b6Ws6KfOjt3RGkqkvrhANt4*3D&reserved=0__;JSUlJSUlJSUlJSUqKiUqKiUqKiUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!SgRmB49jMkpEGwajZUGOeFs1vKGra6qC9bvGoQ_-R5gmjutWm1SkLioAbZS2KOX3gA$>,
La Jolla, CA 92037
***************************





_______________________________________________

3dem mailing list

3dem at ncmir.ucsd.edu<mailto:3dem at ncmir.ucsd.edu>

https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem<https://urldefense.com/v3/__https:/nam11.safelinks.protection.outlook.com/?url=https*3A*2F*2Fmail.ncmir.ucsd.edu*2Fmailman*2Flistinfo*2F3dem&data=02*7C01*7Cmbsherma*40utmb.edu*7C7588ba702e3d48b0331608d82149701a*7C7bef256d85db4526a72d31aea2546852*7C0*7C0*7C637295946590215732&sdata=mdkP3tHxoFaAojUJYWNUoEIOC4R05amSuIN1N2ri8Hk*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUl!!Mih3wA!SgRmB49jMkpEGwajZUGOeFs1vKGra6qC9bvGoQ_-R5gmjutWm1SkLioAbZSqujSnDw$>
_______________________________________________
3dem mailing list
3dem at ncmir.ucsd.edu<mailto:3dem at ncmir.ucsd.edu>
https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem<https://urldefense.com/v3/__https:/nam11.safelinks.protection.outlook.com/?url=https*3A*2F*2Fmail.ncmir.ucsd.edu*2Fmailman*2Flistinfo*2F3dem&data=02*7C01*7Cmbsherma*40utmb.edu*7C7588ba702e3d48b0331608d82149701a*7C7bef256d85db4526a72d31aea2546852*7C0*7C0*7C637295946590215732&sdata=mdkP3tHxoFaAojUJYWNUoEIOC4R05amSuIN1N2ri8Hk*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUl!!Mih3wA!SgRmB49jMkpEGwajZUGOeFs1vKGra6qC9bvGoQ_-R5gmjutWm1SkLioAbZSqujSnDw$>
_______________________________________________
3dem mailing list
3dem at ncmir.ucsd.edu<mailto:3dem at ncmir.ucsd.edu>
https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem<https://urldefense.com/v3/__https:/nam11.safelinks.protection.outlook.com/?url=https*3A*2F*2Fmail.ncmir.ucsd.edu*2Fmailman*2Flistinfo*2F3dem&data=02*7C01*7Cmbsherma*40utmb.edu*7C7588ba702e3d48b0331608d82149701a*7C7bef256d85db4526a72d31aea2546852*7C0*7C0*7C637295946590215732&sdata=mdkP3tHxoFaAojUJYWNUoEIOC4R05amSuIN1N2ri8Hk*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUl!!Mih3wA!SgRmB49jMkpEGwajZUGOeFs1vKGra6qC9bvGoQ_-R5gmjutWm1SkLioAbZSqujSnDw$>



_______________________________________________

3dem mailing list

3dem at ncmir.ucsd.edu<mailto:3dem at ncmir.ucsd.edu>

https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem
_______________________________________________
3dem mailing list
3dem at ncmir.ucsd.edu
https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem

________________________________

Important: This email is confidential and may be privileged. If you are not the intended recipient, please delete it and notify us immediately; you should not copy or use it for any purpose, nor disclose its contents to any other person. Thank you.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://mail.ncmir.ucsd.edu/pipermail/3dem/attachments/20200706/ef2bf111/attachment-0001.html>


More information about the 3dem mailing list