[3dem] [TEM] CTF and SNR

Marin van Heel marin.vanheel at googlemail.com
Fri Jun 9 02:57:18 PDT 2017


Hi Philip,

Admittedly, the wave-particle duality of nature is a confusing matter!  
All wave functions and (complex) transmission functions in the object 
plane and in diffraction plane (back-focal plane of the lens) are only 
probability functions. They only become real entities when they are 
measured in the image plane (the waves squared, are actually measured).  
It is there where we introduce the counting of the arriving electrons 
(particles), a counting which is subject to Poisson statistics.   Thus 
Poisson statistics is NOT subject to the CTF. (See Alexis' post below)

Cheers

Marin

================================================


On 09/06/2017 10:32, Philip Koeck wrote:
>
> Hi again.
>
> Several people have pointed out that shot-noise is the main noise 
> contribution and that it isn’t modulated by the CTF.
>
> I’m having a hard time understanding why it wouldn’t be modulated by 
> the CTF.
>
> All shot noise that occurs before the imaging by the objective lens 
> should somehow affect the exit wave, shouldn’t it?
>
> (I know I’m switching from a particle to a wave description here.)
> Since the whole exit wave (signal and noise contributions) is 
> multiplied by the same phase factor containing the
> lens aberration function, the noise in it should be affected by the 
> CTF just as much as the signal.
> Now I’m thinking about shot noise produced mainly during elastic 
> scattering in the specimen.
>
> I’m trying not to think of inelastic scattering to keep things simpler.
>
> All the best,
>
> Philip
>
> *Från:*3dem [mailto:3dem-bounces at ncmir.ucsd.edu] *För *Alexis Rohou
> *Skickat:* den 7 juni 2017 15:25
> *Till:* 3dem at ncmir.ucsd.edu
> *Ämne:* Re: [3dem] [TEM] CTF and SNR
>
> Hi Philip,
>
>     I’ll assume most of the noise in an image of an ice-embedded
>     bio-molecule is due to variations in the ice, so called structural
>     noise.
>
> This is not a safe assumption. In fact most of the noise is shot 
> noise, which is Poisson distributed (so, Gaussian for our total 
> exposures), and not modulated by the CTF. Hence the CTF does modulate 
> the SSNR.
>
> A few years ago, Joachim Frank and colleagues attempted a 
> quantification of this and other sources of noise. You may find this a 
> good read. They come up with a shot noise SNR of ~ 0.1, and a combined 
> shot+structural noise SNR of ~1, if I read them correctly.
>
> Baxter, W. T., Grassucci, R. A., Gao, H., & Frank, J. (2009). 
> Determination of signal-to-noise ratios and spectral SNRs in cryo-EM 
> low-dose imaging of molecules. Journal of Structural Biology, 166(2), 
> 126–32. https://doi.org/10.1016/j.jsb.2009.02.012
>
> Cheers,
> Alexis
>
> On 06/07/2017 01:22 AM, Philip Koeck wrote:
>
>     Hi.
>
>     I’ve recently encountered a bit of a conundrum:
>
>     I’ll assume most of the noise in an image of an ice-embedded
>     bio-molecule is due to variations in the ice, so called structural
>     noise.
>
>     The CTF describes the contrast transfer for both signal and noise
>     in the same way.
>
>     So, what is the point of changing the CTF by defocusing and/or
>     using a phase plate.
>     The SNR should be unchanged.
>
>     I can think of one explanation: If the signal spectrum is very
>     different from the noise spectrum one could chose a CTF that
>     enhances the
>     resolution bands where the difference is big.
>
>     Else: Is there some other factor that affects visibility of the
>     molecule than SNR?
>
>     All the best,
>
>     Philip
>
>
>
>
>     _______________________________________________
>
>     3dem mailing list
>
>     3dem at ncmir.ucsd.edu <mailto:3dem at ncmir.ucsd.edu>
>
>     https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem
>
>
>
> _______________________________________________
> 3dem mailing list
> 3dem at ncmir.ucsd.edu
> https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem


-- 
==============================================================

     Prof Dr Ir Marin van Heel

     Research Professor at:

     Laboratório Nacional de Nanotecnologia - LNNano
     CNPEM/ABTLuS, Campinas, Brazil
  

-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://mail.ncmir.ucsd.edu/pipermail/3dem/attachments/20170609/ff3df2de/attachment-0001.html>


More information about the 3dem mailing list