Hi,

Thanks for your suggestions. Surely I would convey your suggestions to the
FEI engineer who has come to check up. Unfortunately most of their service
engineers (at least the ones sent to our lab) are not that much acquainted
with Polara. They had been just checking what the problem might be with
different trials and came to the end of their stipulated time of visit and
hand over the problem to their indian partners who are not even trained
with polara. Hence I have been looking for these suggestions and will look
forward more so that I can at least suggest them the ways to fix the
problem. I have also seen that there are many of the FEI persons over here.
I would be highly obliged if any expert from them can put forward some
points.

Sayan

On Wed, Sep 7, 2016 at 12:26 AM, Guenter Resch <lists at nexperion.net> wrote:

> Hi,
>
> > Could also additionally be a temperature problem when e.g. the objective
> aperture holder touches the cryo-box, also have this checked by service
> engineer.
> > I've also had issues here from users who did not clip the grids
> properly. Without the clip-ring fully pushing the grid into the cartridge,
> contamination was horrible after just one night.
>
> Just for the sake of completeness, one more reason for why the temperature
> of the sample in a Polara can be too high: if the flexible cooling braids
> (bundles of thin copper fibres) connecting the static cryo boxes to the
> shifting and tilting counter-piece where the cartridge sits are worn out
> (i.e., individual copper fibres are broken), the cooling efficiency is
> lowered by the diminished diameter. I had this once, resulting in more or
> less fast deterioration of sample quality (leopard skin effect).
>
> In addition to the problems Wim mentioned, this should be checked as well
> if the column is split (or even via an endoscope).
>
> Best,
>
>         Guenter
>
> --
> Dr. Guenter Resch
> Nexperion e.U. - Solutions for Electron Microscopy - www.nexperion.net
> Mattiellistrasse 3/17, 1040 Vienna, Austria - Phone +43 664 94 17 210
> Registered at Commercial Court Vienna, FN 397677w - VAT Reg. ATU67962234
>



-- 

Sayan Bhakta

Research Scholar

Group of Dr. Jayati Sengupta

CSIR - Indian Institute of Chemical Biology

Kolkata, West Bengal

India