<div dir="ltr"><div>Hi,</div><div><br><div>Thanks for your suggestions. Surely I would convey your suggestions to the FEI engineer who has come to check up. Unfortunately most of their service engineers (at least the ones sent to our lab) are not that much acquainted with Polara. They had been just checking what the problem might be with different trials and came to the end of their stipulated time of visit and hand over the problem to their indian partners who are not even trained with polara. Hence I have been looking for these suggestions and will look forward more so that I can at least suggest them the ways to fix the problem. I have also seen that there are many of the FEI persons over here. I would be highly obliged if any expert from them can put forward some points.</div></div><div><br></div><div>Sayan</div></div><div class="gmail_extra"><br><div class="gmail_quote">On Wed, Sep 7, 2016 at 12:26 AM, Guenter Resch <span dir="ltr"><<a href="mailto:lists@nexperion.net" target="_blank">lists@nexperion.net</a>></span> wrote:<br><blockquote class="gmail_quote" style="margin:0 0 0 .8ex;border-left:1px #ccc solid;padding-left:1ex">Hi,<br>
<span class=""><br>
> Could also additionally be a temperature problem when e.g. the objective aperture holder touches the cryo-box, also have this checked by service engineer.<br>
> I've also had issues here from users who did not clip the grids properly. Without the clip-ring fully pushing the grid into the cartridge, contamination was horrible after just one night.<br>
<br>
</span>Just for the sake of completeness, one more reason for why the temperature of the sample in a Polara can be too high: if the flexible cooling braids (bundles of thin copper fibres) connecting the static cryo boxes to the shifting and tilting counter-piece where the cartridge sits are worn out (i.e., individual copper fibres are broken), the cooling efficiency is lowered by the diminished diameter. I had this once, resulting in more or less fast deterioration of sample quality (leopard skin effect).<br>
<br>
In addition to the problems Wim mentioned, this should be checked as well if the column is split (or even via an endoscope).<br>
<br>
Best,<br>
<br>
        Guenter<br>
<span class="HOEnZb"><font color="#888888"><br>
--<br>
Dr. Guenter Resch<br>
Nexperion e.U. - Solutions for Electron Microscopy - <a href="http://www.nexperion.net" rel="noreferrer" target="_blank">www.nexperion.net</a><br>
Mattiellistrasse 3/17, 1040 Vienna, Austria - Phone +43 664 94 17 210<br>
Registered at Commercial Court Vienna, FN 397677w - VAT Reg. ATU67962234<br>
</font></span></blockquote></div><br><br clear="all"><div><br></div>-- <br><div class="gmail_signature" data-smartmail="gmail_signature"><div dir="ltr"><div><div dir="ltr"><p style="font-size:12.8px;margin:0cm 0cm 0.0001pt;background-image:initial;background-repeat:initial"><span style="color:black"><font face="arial, helvetica, sans-serif">Sayan Bhakta</font></span></p><p style="font-size:12.8px;margin:0cm 0cm 0.0001pt;background-image:initial;background-repeat:initial"><span style="color:black"><font face="arial, helvetica, sans-serif">Research Scholar</font></span></p><p style="font-size:12.8px;margin:0cm 0cm 0.0001pt;background-image:initial;background-repeat:initial"><span style="color:black"><font face="arial, helvetica, sans-serif">Group of Dr. Jayati Sengupta</font></span></p><p style="font-size:12.8px;margin:0cm 0cm 0.0001pt;background-image:initial;background-repeat:initial"><span style="color:black"><font face="arial, helvetica, sans-serif">CSIR - Indian Institute of Chemical Biology</font></span></p><p style="font-size:12.8px;margin:0cm 0cm 0.0001pt;background-image:initial;background-repeat:initial"><span style="color:black"><font face="arial, helvetica, sans-serif">Kolkata, West Bengal</font></span></p><p style="font-size:12.8px;margin:0cm 0cm 0.0001pt;background-image:initial;background-repeat:initial"><span style="color:black"><font face="arial, helvetica, sans-serif">India</font></span></p></div></div></div></div>
</div>