Dear Roman and Benjamin,

I don’t understand your comments about Friedel Symmetry.   In my understanding, Friedel Symmetry is referring to the fact that any Fourier voxel (h,k,l) with values (AMP,PHA) has a Friedel-related voxel (-h,-k,-l) on the other side of the 3D Fourier space, which is complex conjugated to the first one, meaning (AMP,-PHA).   As this usually applies to Fourier space, most software packages only save one half of the Fourier volume, since the other half is anyway the same, except the complex conjugation.

There are very rare cases, where Friedel symmetry does not uphold, such as in dynamical electron diffraction, but these are probably not what you refer to.

So, what do you refer to by saying the below?

Benjamin:
3) Neither presentation seemed to reflect friedel symmetry. While your choice of cone axis could result in asymmetric sampling, I find this curious. Comments?
Roman:
The cones are chosen such that they cover the whole Fourier Space but in opposite directions they are not necessarily exactly the same (there is no symmetry in the cone distribution). So the choice of cone angles is not symmetrical.

Or, do you refer to another type of symmetry?

Henning.



Henning Stahlberg, PhD
Prof. for Structural Biology, C-CINA, Biozentrum, University Basel
Mattenstrasse 26 | D-BSSE | WRO-1058 | CH-4058 Basel | Switzerland
http://c-cina.org | Tel. +41-61-387 32 62



On Sep 7, 2015, at 11:30, R.I.Koning at lumc.nl<mailto:R.I.Koning at lumc.nl> wrote:

Dear Benjamin,

Different spatial sampling, including under sampling, is definitely an issue. I am not sure what you mean with concern, but it is something to keep in mind. You have to understand the limiting factor in the data and the measurements in order to not over interpret any data. Doing concial FSC on tomographic data is definitely not a perfect “resolution” measurement criterion (is there is any).  I would not put any emphasis on the actual values that originate from the individual cones. However, I hope we showed that can be useful when one want to compare similar data or reconstruction techniques in a relative and qualitative manner.

We did try other than 200 cones and this number is optimised for spatial distribution against signal to noise.

Interesting though to scale the resolution relatively. I am not sure I understand what would be 1 then. It probably does not matter since the power is in the relative values anyway. Only maybe when one has a subtomogram average with more data than from one tomogram the values might be somewhat more meaningful.

I am not sure if I understand you last remark. The cones are chosen such that they cover the whole Fourier Space but in opposite directions they are not necessarily exactly the same (there is no symmetry in the cone distribution). So the choice of cone angles is not symmetrical.

Best regards,

Roman


Dr. R.I.Koning (Roman)

Department of Cell Biology, Leiden University Medical Center
Postal Zone S1-P, P.O.Box 9600, 2300 RC Leiden, The Netherlands
visiting address: Einthovenweg 20, Building 2, zone S1-P, room R-90-20
tel. (31) 71 526 9296 fax. (+31) 71 526 8270 mail. r.i.koning at lumc.nl<mailto:r.i.koning at lumc.nl> web. electronmicroscopy.lumc.nl<http://electronmicroscopy.lumc.nl>

Netherlands Centre for Electron Nanoscopy, Institute Biology Leiden, Leiden University
visiting address: Einsteinweg 55, 2333 RC Leiden, The Netherlands, room 05.11
tel. (31) 71 527 1423 mail. r.i.koning at biology.leidenuniv.nl<mailto:r.i.koning at biology.leidenuniv.nl> web. www.necen.nl<http://www.necen.nl>

From: Benjamin Himes <himes.benjamin at gmail.com<mailto:himes.benjamin at gmail.com>>
Date: Sunday 6 September 2015 15:30
To: Roman Koning <r.i.koning at lumc.nl<mailto:r.i.koning at lumc.nl>>
Cc: "3dem at ncmir.ucsd.edu<mailto:3dem at ncmir.ucsd.edu>" <3dem at ncmir.ucsd.edu<mailto:3dem at ncmir.ucsd.edu>>
Subject: RE: [3dem] Anisotropic resolution

Dear Roman & Misha,
Thank you both for pointing out your respective papers regarding the subdivision of Fourier space by shells & cones in order to estimate resolution anisotropy.
Follow up questions:
1) are you concerned with the effect of undersampling an already sparse transform (particularly in the case of the tomogram work?)
Roman - you report use of 200 cones. Did you try other numbers, and what did you observe; I would guess the overall pattern to be very similar with different "resolution" values?
2) both papers reference resolution comparing in either pixel^-1 or nm^-1. Depending on your answer to question 1, do you think this is appropriate, or would a relative measure (scale the highest to 1) perhaps be a better representation.
3) Neither presentation seemed to reflect friedel symmetry. While your choice of cone axis could result in asymmetric sampling, I find this curious. Comments?
Thanks again!
Benjamin Himes

Dear Benjamin,

We recently published a paper on how to measure the anisotropy in electron tomograms (http://www.ncbi.nlm.nih.gov/pubmed/25843950). Though we did not test this on subtomogram averages it might be useful to have a look at the paper (in case you did not do that yet).

Best regards,

Roman

Dr. R.I.Koning (Roman)

Department of Cell Biology, Leiden University Medical Center
Postal Zone S1-P, P.O.Box 9600, 2300 RC Leiden, The Netherlands
visiting address: Einthovenweg 20, Building 2, zone S1-P, room R-90-20
tel. (31) 71 526 9296 fax. (+31) 71 526 8270<tel:%28%2B31%29%2071%20526%208270> mail. r.i.koning at lumc.nl<mailto:r.i.koning at lumc.nl> web. electronmicroscopy.lumc.nl<http://electronmicroscopy.lumc.nl/>

Netherlands Centre for Electron Nanoscopy, Institute Biology Leiden, Leiden University
visiting address: Einsteinweg 55, 2333 RC Leiden, The Netherlands, room 05.11
tel. (31) 71 527 1423 mail. r.i.koning at biology.leidenuniv.nl<mailto:r.i.koning at biology.leidenuniv.nl> web. www.necen.nl<http://www.necen.nl/>


From: 3dem [mailto:3dem-bounces at ncmir.ucsd.edu<mailto:3dem-bounces at ncmir.ucsd.edu>] On Behalf Of Benjamin Himes
Sent: maandag 24 augustus 2015 15:44
To: 3dem at ncmir.ucsd.edu<mailto:3dem at ncmir.ucsd.edu>
Subject: [3dem] Anisotropic resolution

Dear colleagues,
I hate to stir the pot on the resolution discussion, however, I have a question regarding any progress on analyzing the anisotropy potentially present in a 3d em map.
Aside from some discussion on 3d ssnr, particularly by P.P. and a recent paper on 3d covariance estimation from J.F. Has anybody tried to implement a concrete way of assessing this issue?
I am particularly concerned with maps generated from subTomogram averaging and classification: even with an angular distribution that is well sampled, I am concerned about the fact that cumulative dose, and increased sample thickness on tilting create a situation where the individual projections do not have equivalent SNR and therefore a simple plot of angular distributions would not accurately reflect the quality of the sampling in Fourier space.
It seems a 3d SSNR might work, but the interpretation of such a plot, beyond its "potato" like quality, even in terms of the eigenvalues of the principle axes is not immediately clear to me.
Many thanks,
Benjamin Himes
_______________________________________________
3dem mailing list
3dem at ncmir.ucsd.edu<mailto:3dem at ncmir.ucsd.edu>
https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem

-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://mail.ncmir.ucsd.edu/pipermail/3dem/attachments/20150908/25aa381a/attachment-0001.html>


More information about the 3dem mailing list