Dear Frank,
We had this problem at Purdue as well and Paul was able to figure it out.
Is the body of the workstation frosting up where it joins the shaft?  If so
it, the O ring has worn out.  There is an O ring at the joint between the
shaft and the cup.  Replace it and you should be back to normal.

Cheers,

Valorie


Valorie Bowman 
Senior Research Electron Microscopist,

Dept. of Biology
Purdue Univ.
W. Lafayette, In 
    47906


 ph. 765-494-5643
 fax 765-496-1189
vdb at purdue.edu


> From: Frank Polzer <frank.polzer at physik.hu-berlin.de>
> Date: Mon, 25 Feb 2013 12:44:11 +0100
> To: <3dem at ncmir.ucsd.edu>
> Subject: [3dem] Gatan 914 cryotransfer holder
> 
> Dear Colleagues,
> 
>  
> 
> I have little trouble with our Gatan 914 cryo transfer holder getting stuck
> in the workstation after cooling it down.
> 
>  
> 
> Has anyone experienced similar problems and knows how to get rid of that
> problem? Of course I try to avoid any remains of condensed water in the
> workstation before cooling it with LN2.
> 
>  
> 
> Thank you all in advance for your help.
> 
>  
> 
> Cheers,
> Frank
> 
>  
> 
> Dr. Frank Polzer 
> TEM Group 
> Institute of Physics
> Humboldt Universität zu Berlin
> Newtonstraße 15 
> 12489 - Berlin 
>   
> Tel.: +49 30 2093-4995 (office)
> Tel.: +49 30 2093-7829 (TEM)
> Fax: +49 30 2093-7760
>   
>  <mailto:frank.polzer at physik.hu-berlin.de> frank.polzer at physik.hu-berlin.de
> 
>  
> 
> _______________________________________________
> 3dem mailing list
> 3dem at ncmir.ucsd.edu
> https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem



More information about the 3dem mailing list