The air bubbles are a real problem. I tried to use a wire as a syringe or mount the tube to a P10 pipette but still saw holes in the ice. Any thoughts about how to avoid those bubbles?
Thanks,
Chen
On Feb 14, 2013, at 1:20 PM, John Crum wrote:
> No. The ends can remain open. Just make sure the tube is filled with liquid. Any air will prevent adequate freezing.
>
> jc
>
>
>
> On 2/13/13 9:16 PM, Chen Chen wrote:
>> Dear Lists,
>>
>> Has anyone tried to high pressure freeze samples in a tube using Leica HPM100? I am wondering if I need to seal one or both ends of the tube before freezing. Thanks in advance.
>>
>> Cheers,