[3dem] Vitrobot - plunging mechanism

Bob Grassucci rg2502 at columbia.edu
Tue Jul 29 06:04:02 PDT 2008


Hi Oren,
If you look  at the blotting pads there is a metal wire that engages 
teeth in the back of the blotting pads that cause the pads to rotate 
after each blot.  My guess is that that pin is either bent or pushed to 
the side so it does not rotate.  Have a look behind the pads.
Regards,
Bob

Oren Regev wrote:
>
> Dear all,
>
> We use Vitrobot (first commercial FEI version) and have serious problems
>
> with the blotting mechanism unit: blotting is not completed and the 
> filter papers do not rotate.
>
> Have any of you share the same problems, or solved this problem?
>
> Thank you,
>
> ----------------------------------
>
> Oren Regev
>
> Dept. Chemical Engineering
>
> Ben-Gurion University of the Negev
>
> 84105  Beer-Sheva
>
> email: oregev at bgu.ac.il
>
> http://www.bgu.ac.il/RS_Minerva/CV_Oren_regev.htm
>
> Tel +972 8 6472145
>
> ------------------------------------------------------------------------
>
> _______________________________________________
> 3dem mailing list
> 3dem at ncmir.ucsd.edu
> https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem
>   


-- 
*********************************
Howard Hughes Medical Institute
Columbia University Medical Center
Black Building Room 2-221
630 West 168th Street
New York, NY 10032-3702
rg2502 at columbia.edu
Phone: 212-305-9520
Fax: 212-305-9500



More information about the 3dem mailing list