Hi David,
There's been a lot of excellent advice already, but I thought I'd add my
experience.
If you want to measure the Cs, the magnification/pixel size has to be
accurately calibrated first. There are several ways of doing this that work
pretty well. I like to use the lattice spacing in a thin gold foil, such as
a HexAuFoil grid, since you can calibrate all the high mags accurately with
just a few images per magnification. There is a program to do this and a
paper comparing several pixel size calibration methods here:
https://urldefense.com/v3/__https://www.mrc-lmb.cam.ac.uk/crusso/magCalEM/index.html__;!!Mih3wA!F7gjnTy38fid9HhbLpIbXtKJirVZwehFZ7mOBOXX0rZBHNlZGlSYabbxOHTGf916JKfm4QRvX3JSvs0IOEYpEZCBElw$
Once you have the pixel size, you can use CTF refinement as others have
suggested. If you also calibrate your beam-tilt coils, which can also be
done using gold, you can also measure the Cs using beam-tilt-induced image
shifts in a method described here
https://urldefense.com/v3/__https://www.sciencedirect.com/science/article/pii/0304399191901583__;!!Mih3wA!F7gjnTy38fid9HhbLpIbXtKJirVZwehFZ7mOBOXX0rZBHNlZGlSYabbxOHTGf916JKfm4QRvX3JSvs0IOEYpUkG2rFk$
>From my experience, the manufacturer provided Cs value has always been
accurate, and apparent errors have been pixel size calibration errors.
Best wishes,
Josh
On Mon, Feb 16, 2026 at 1:07 PM Morgan, David Gene via 3dem <
3dem at ncmir.ucsd.edu> wrote:
> Hi,
>
> For those of you who have used any of the image processing programs to
> refine the Cs value for your microscope, how does the refined value compare
> to what the manufacturers claim?
>
> As I understand it, the manufacturers simply calculate a Cs value for each
> microscope/pole piece model. I guess I have always assumed that those
> values are relatively accurate, but I don't really have any data to support
> that. Nor am I certain what would constitute "relatively accurate."
>
> Any thoughts would be appreciated. Thanks.
>
> --
>
> NOTE: my email has changed to dagmorga at iu.edu.
> Email to dagmorga at indiana.edu will stop being
> forwarded later this year.
>
> politics is more difficult than physics.
> A. Einstein
>
> David Gene Morgan
> Electron Microscopy Center
> 043E Simon Hall
> IU Bloomington
> 812 856 1457 (office)
> 812 856 3221 (3200)
> https://urldefense.com/v3/__http://iubemcenter.indiana.edu__;!!Mih3wA!F7gjnTy38fid9HhbLpIbXtKJirVZwehFZ7mOBOXX0rZBHNlZGlSYabbxOHTGf916JKfm4QRvX3JSvs0IOEYpexi88h4$
> <https://urldefense.com/v3/__http://iubemcenter.indiana.edu__;!!Mih3wA!GtPRGTEs6TwC_HvS-grWkxuNLZZrTPkNnxibfaqE1R2wuugqzQRLCXxNTJ6tcwkHZ9lKP0Ewa1fN0w9zI0Y$>
> _______________________________________________
> 3dem mailing list
> 3dem at ncmir.ucsd.edu
> https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem
>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://mail.ncmir.ucsd.edu/pipermail/3dem/attachments/20260217/1a57a82d/attachment.html>
More information about the 3dem
mailing list