[3dem] How to clean the Aquilos 2 specimen shuttle
Paolo Swuec
paolo.swuec at gmail.com
Fri Feb 2 12:10:44 PST 2024
IPA and a brush will do the trick! Rinse well with DDI water and dry properly, we have never have issue cleaning like this and they go back to native state.
Paolo
> Il giorno 2 feb 2024, alle ore 20:20, Zongli Li <zongli.li at gmail.com> ha scritto:
>
>
> Dear colleagues,
>
> After many times uses with multiple sputter coatings and GIS depositions, our 35/45 degree pre-tilt specimen shutters became very dirty and do not hold the auto-grids well. We are asking communities for ways to clean them up or restore their functionality. Any ideas or recommendations would be highly appreciated.
>
> Thank you,
>
> Best regards,
>
> Zongli