[3dem] [+externe Mail+] Nitrogen support for Aquilos2
Jean Daraspe
jean.daraspe at unil.ch
Thu Jul 6 08:42:44 PDT 2023
Dear Xinran,
The N2 flow needed to cool down the Aquilos 2 stage is 9L/min (not 30 liters/min)
It means that with a dewar of 250 liters of liquid nitrogen you can cool it for 13 days.
The problem we have at the beginning is that after 12-14 hours of running the Aquilos the blue tubing of N2 entrance was plugged by ice contamination. A very low amount of water in the N2 is enough to plug these small tube: A column of 200µl of water can plug this tube (even less). It means that 0.02ppm of water in your N2 is enough to plug it after 14hours.
Two possibilities:
- You have the PD-2012 Portable Rare Gas Dryer from Sircal Instruments (UK) that can dry your gas and cost around 2000€. It's a small red box that work with molecular sieve, and need regeneration after some days (just switch on the box for 4 hours at the end of the week and switch it off, it's enough). But even with it after two days my blue tube is plugged with ice. It's better than nothing but not perfect. Maybe I didn't regenerate the sieve enough...I don't know.
- That's why I install a passive cold trap: it's cheap and very efficient (see attachment), that's the best in my opinion. Just put the longer tube as you can in the cooling dewar. Mount it at the top of the cooling head, just below the black cap. The longer is the better to trap all the water from the N2. Now, after 4 days of continuous N2 flow in this blue tube, there is no plug. Of course you need a "cryo cycle" at the end of your long run to dry the whole system.
Best,
Jean
--------------------------------------------------
Jean Daraspe, MSc | Lab Scientist
UNIL | Université de Lausanne
Faculté de biologie et de médecine
Electron Microscopy Facility
Bâtiment Biophore - Quartier Sorge
Room 1217
CH-1015 Lausanne
Tél: +41216925061
<mailto:irina.kolotueva at unil.ch>jean.daraspe at unil.ch<mailto:jean.daraspe at unil.ch>
https://urldefense.com/v3/__http://wp.unil.ch/emf__;!!Mih3wA!DWcQkg0Ty4cqq9C59GJZt1mAPqlCBYBOcOpahwXVeCJqSDle1NQC0O0UXr0ojcgeckgwRADuGyu_zIMjgxAUzoILbw$
________________________________
De : 3dem <3dem-bounces at ncmir.ucsd.edu> de la part de Cheng, Tat Cheung <tatcheung.cheng at med.uni-goettingen.de>
Envoyé : jeudi 6 juillet 2023 15:13
À : 3dem at ncmir.ucsd.edu <3dem at ncmir.ucsd.edu>
Objet : Re: [3dem] [+externe Mail+] Nitrogen support for Aquilos2
Hi Chunyan,
It doesn't cost much since it is just liquid nitrogen turned gas, if you mean by nitrogen cost. But if you mean by running fee of the system, then it varies between places.
Tat
On 06.07.23 14:39, Wang, Chunyan wrote:
Hi Tat, that sounds great to be able to run Aquilos2 for multiple days. How much does that cost more or less ?
Best
Chunyan Wang
Microbial Sciences Institute
Department of Microbial Pathogenesis
Yale School of Medicine
840 West Campus Drive
New Haven, CT, 06516
________________________________
From: 3dem <3dem-bounces at ncmir.ucsd.edu><mailto:3dem-bounces at ncmir.ucsd.edu> on behalf of 3dem-request at ncmir.ucsd.edu<mailto:3dem-request at ncmir.ucsd.edu> <3dem-request at ncmir.ucsd.edu><mailto:3dem-request at ncmir.ucsd.edu>
Sent: Thursday, July 6, 2023 4:16 AM
To: 3dem at ncmir.ucsd.edu<mailto:3dem at ncmir.ucsd.edu> <3dem at ncmir.ucsd.edu><mailto:3dem at ncmir.ucsd.edu>
Subject: 3dem Digest, Vol 191, Issue 7
Send 3dem mailing list submissions to
3dem at ncmir.ucsd.edu<mailto:3dem at ncmir.ucsd.edu>
To subscribe or unsubscribe via the World Wide Web, visit
https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Fmail.ncmir.ucsd.edu*2Fmailman*2Flistinfo*2F3dem&data=05*7C01*7Cchunyan.wang.cw785*40yale.edu*7C91bc1a7b767a4e026de808db7df95db7*7Cdd8cbebb21394df8b4114e3e87abeb5c*7C0*7C0*7C638242282236345923*7CUnknown*7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0*3D*7C3000*7C*7C*7C&sdata=LBKEC6KGa5N*2ByMLYN*2FRucvcRmwjcAE8Pzu11gnHxWtg*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUl!!Mih3wA!DWcQkg0Ty4cqq9C59GJZt1mAPqlCBYBOcOpahwXVeCJqSDle1NQC0O0UXr0ojcgeckgwRADuGyu_zIMjgxCxPWCAMQ$ <https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem>
or, via email, send a message with subject or body 'help' to
3dem-request at ncmir.ucsd.edu<mailto:3dem-request at ncmir.ucsd.edu>
You can reach the person managing the list at
3dem-owner at ncmir.ucsd.edu<mailto:3dem-owner at ncmir.ucsd.edu>
When replying, please edit your Subject line so it is more specific
than "Re: Contents of 3dem digest..."
Today's Topics:
1. Re: [+externe Mail+] Re: Dry nitrogen generator that supports
Aquilos cryo FIB (Cheng, Tat Cheung)
2. DC EMI cancellation (Hannah Turner)
3. Going from Amira filament picks to subtomogram averaging
(Matthew Chang)
4. Re: DC EMI cancellation (Andrzej ?ak)
5. Post-doc in Paris (Kikuti Carlos)
6. Re: DC EMI cancellation (Stahlberg Henning)
----------------------------------------------------------------------
Message: 1
Date: Wed, 5 Jul 2023 19:40:39 +0000
From: "Cheng, Tat Cheung" <tatcheung.cheng at med.uni-goettingen.de><mailto:tatcheung.cheng at med.uni-goettingen.de>
To: "3dem at ncmir.ucsd.edu"<mailto:3dem at ncmir.ucsd.edu> <3dem at ncmir.ucsd.edu><mailto:3dem at ncmir.ucsd.edu>
Subject: Re: [3dem] [+externe Mail+] Re: Dry nitrogen generator that
supports Aquilos cryo FIB
Message-ID:
<d5f7acec-eb6c-0621-dfca-fbb6a3cb2626 at med.uni-goettingen.de><mailto:d5f7acec-eb6c-0621-dfca-fbb6a3cb2626 at med.uni-goettingen.de>
Content-Type: text/plain; charset="utf-8"
Hi Xinran,
We also don't have house N2 gas. What we do is to use the gas outlet from a Taylor-Wharton 250L liquid nitrogen tank, with a higher pressure regulator (set to 6 bar max). It is not a standard usage of the tank, but it works for us. One thing to note is the cold N2 gas outlet needs to be warmed up to prevent crazy condensation. One tank lasts more than 1 week. While it may be hard to measure the water content in the N2 gas outlet, and it may depend on the cleanliness of the liquid N2, but our setup allows continuous usage of the Aquilos2 for multiple days.
Tat
Multiscale Bioimaging Cluster of Excellence
Fakult?t der Physik, D-1.111
Friedrich-Hund-Platz 1, 37077 G?ttingen
Telephon: +49 (0)551 / 39-60783
https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Fmbexc.de*2F__*3B!!Mih3wA!G_yCkpuJFHFFZBEfmx9IUF6C-pAF4D5BYPK9R5bhg8DGZAavOISpoNqluoQaME1RBvM013mrNAlGpHUYz1CY9i8s6YJkW2xxV1CmM6eDvQ*24&data=05*7C01*7Cchunyan.wang.cw785*40yale.edu*7C91bc1a7b767a4e026de808db7df95db7*7Cdd8cbebb21394df8b4114e3e87abeb5c*7C0*7C0*7C638242282236502146*7CUnknown*7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0*3D*7C3000*7C*7C*7C&sdata=0WMR84NNypuERjizoyhWGO*2FioEi*2Fl2*2F7wnHAfIv0A0Q*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUl!!Mih3wA!DWcQkg0Ty4cqq9C59GJZt1mAPqlCBYBOcOpahwXVeCJqSDle1NQC0O0UXr0ojcgeckgwRADuGyu_zIMjgxCbjHK4mA$ <https://urldefense.com/v3/__https://mbexc.de/__;!!Mih3wA!G_yCkpuJFHFFZBEfmx9IUF6C-pAF4D5BYPK9R5bhg8DGZAavOISpoNqluoQaME1RBvM013mrNAlGpHUYz1CY9i8s6YJkW2xxV1CmM6eDvQ$>
On 05.07.23 20:54, Sasha Dickinson wrote:
Hi Xinran,
We also do not have house N2 gas so I tried a few different N2 sources and found that the best / most reasonable one was a standard ~5000 cubic feet tank at 350 PSI (as opposed to the typical 22 PSI). This tank usually lasts us >1 week on our Aquilos 2, so I usually have it replaced weekly like the ones for the cryoTEMs.
Very best,
Sasha
On Wed, Jul 5, 2023 at 11:33?AM Liu, Xinran <xinran.liu at yale.edu<mailto:xinran.liu at yale.edu><mailto:xinran.liu at yale.edu><mailto:xinran.liu at yale.edu>> wrote:
Dear Colleagues,
I am looking for help to identify a dry N2 generator that can work with the Aquilos II cryo FIB (Thermo Fisher).
It has become a concern that the Aquilos milling operation requires a huge volume of dry nitrogen that makes use of typical cylinders not practical. Because our building does not have a N2 supply source, we will have to purchase and install one that works with the Aquilos. Our multiple attempts to reach out to Thermo Fisher have not generated any response so far. The following are requirements for the generator, if someone here knows one or uses one that meet the criteria, can you share the information? much appreciated.
N2 at connection of the microscope
0.5x105 Pa (7.25 psi, 0.5 bar) to 0.8 x105 Pa (11.6 psi, 0.8 bar)
Flow rate is 30 liters/min
The water content must be less than 10 ppm
Xinran Nick Liu, M.D. & Ph.D.
Director, Center for Cellular & Molecular Imaging
Electron Microscopy Core Facility
Senior Research Scientist in Cell Biology
Yale University School of Medicine
Office Phone: 203.785.4050
Lab Phone: 203.785.5390
https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__http*3A*2F*2Fmedicine.yale.edu*2Fccmi*2Fem__*3B!!Mih3wA!G_yCkpuJFHFFZBEfmx9IUF6C-pAF4D5BYPK9R5bhg8DGZAavOISpoNqluoQaME1RBvM013mrNAlGpHUYz1CY9i8s6YJkW2xxV1C7g5TEEg*24&data=05*7C01*7Cchunyan.wang.cw785*40yale.edu*7C91bc1a7b767a4e026de808db7df95db7*7Cdd8cbebb21394df8b4114e3e87abeb5c*7C0*7C0*7C638242282236502146*7CUnknown*7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0*3D*7C3000*7C*7C*7C&sdata=c5Wj0UcdmW2BSv1BZBnriNqSxleiiv7PODfDzDQJa0M*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!DWcQkg0Ty4cqq9C59GJZt1mAPqlCBYBOcOpahwXVeCJqSDle1NQC0O0UXr0ojcgeckgwRADuGyu_zIMjgxBzJbPOlA$ <https://urldefense.com/v3/__http://medicine.yale.edu/ccmi/em__;!!Mih3wA!G_yCkpuJFHFFZBEfmx9IUF6C-pAF4D5BYPK9R5bhg8DGZAavOISpoNqluoQaME1RBvM013mrNAlGpHUYz1CY9i8s6YJkW2xxV1C7g5TEEg$> <https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__http*3A*2F*2Fmedicine.yale.edu*2Fccmi*2Fem__*3B!!Mih3wA!AtSccyzvx8vIvJoc8dL2wPoyPez_WaSgRS0BsBwF70PkXgm7AkzRvmhy_5iwYe0_SR5vmzzTc-8Hlrsd25uk_dz9*24&data=05*7C01*7Cchunyan.wang.cw785*40yale.edu*7C91bc1a7b767a4e026de808db7df95db7*7Cdd8cbebb21394df8b4114e3e87abeb5c*7C0*7C0*7C638242282236502146*7CUnknown*7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0*3D*7C3000*7C*7C*7C&sdata=XlQNOX7H8uiMwUIxiVqljbuGQA1T09mt2xYLOMeWMaA*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!DWcQkg0Ty4cqq9C59GJZt1mAPqlCBYBOcOpahwXVeCJqSDle1NQC0O0UXr0ojcgeckgwRADuGyu_zIMjgxDWbW1MkA$ <https://urldefense.com/v3/__http://medicine.yale.edu/ccmi/em__;!!Mih3wA!AtSccyzvx8vIvJoc8dL2wPoyPez_WaSgRS0BsBwF70PkXgm7AkzRvmhy_5iwYe0_SR5vmzzTc-8Hlrsd25uk_dz9$>>
_______________________________________________
3dem mailing list
3dem at ncmir.ucsd.edu<mailto:3dem at ncmir.ucsd.edu><mailto:3dem at ncmir.ucsd.edu>
https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Fmail.ncmir.ucsd.edu*2Fmailman*2Flistinfo*2F3dem__*3B!!K-Hz7m0Vt54!jdcq28xHmU7qDg3isBVp_HbmV0BDKe5CQzT8quF2QQz3g48b619DGO0O7Mtcmc2FtF2_meELg4JkRZlYcky4*24&data=05*7C01*7Cchunyan.wang.cw785*40yale.edu*7C91bc1a7b767a4e026de808db7df95db7*7Cdd8cbebb21394df8b4114e3e87abeb5c*7C0*7C0*7C638242282236502146*7CUnknown*7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0*3D*7C3000*7C*7C*7C&sdata=lfIWdD38BRztLVP0UUGIR6Y24agAIY6DuUXhxZx58xw*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSU!!Mih3wA!DWcQkg0Ty4cqq9C59GJZt1mAPqlCBYBOcOpahwXVeCJqSDle1NQC0O0UXr0ojcgeckgwRADuGyu_zIMjgxB5j2kLiw$ <https://urldefense.com/v3/__https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem__;!!K-Hz7m0Vt54!jdcq28xHmU7qDg3isBVp_HbmV0BDKe5CQzT8quF2QQz3g48b619DGO0O7Mtcmc2FtF2_meELg4JkRZlYcky4$>
_______________________________________________
3dem mailing list
3dem at ncmir.ucsd.edu<mailto:3dem at ncmir.ucsd.edu><mailto:3dem at ncmir.ucsd.edu>
https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Fmail.ncmir.ucsd.edu*2Fmailman*2Flistinfo*2F3dem&data=05*7C01*7Cchunyan.wang.cw785*40yale.edu*7C91bc1a7b767a4e026de808db7df95db7*7Cdd8cbebb21394df8b4114e3e87abeb5c*7C0*7C0*7C638242282236502146*7CUnknown*7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0*3D*7C3000*7C*7C*7C&sdata=aCQ*2F43X3rSG0Ud493IAAAvHMKzpB*2Fg6*2BOXadE4gB6U4*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!DWcQkg0Ty4cqq9C59GJZt1mAPqlCBYBOcOpahwXVeCJqSDle1NQC0O0UXr0ojcgeckgwRADuGyu_zIMjgxCR4oLwpQ$ <https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=http*3A*2F*2Fmail.ncmir.ucsd.edu*2Fpipermail*2F3dem*2Fattachments*2F20230705*2F3c7c4ea8*2Fattachment-0001.html&data=05*7C01*7Cchunyan.wang.cw785*40yale.edu*7C91bc1a7b767a4e026de808db7df95db7*7Cdd8cbebb21394df8b4114e3e87abeb5c*7C0*7C0*7C638242282236502146*7CUnknown*7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0*3D*7C3000*7C*7C*7C&sdata=3GkTce6Ks2DnufiOhebHTF*2FLn1tkyyR1gRfr4ujWprI*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSU!!Mih3wA!DWcQkg0Ty4cqq9C59GJZt1mAPqlCBYBOcOpahwXVeCJqSDle1NQC0O0UXr0ojcgeckgwRADuGyu_zIMjgxCIu1NWOQ$ <http://mail.ncmir.ucsd.edu/pipermail/3dem/attachments/20230705/3c7c4ea8/attachment-0001.html>>
------------------------------
Message: 2
Date: Wed, 5 Jul 2023 20:05:54 +0000
From: Hannah Turner <hannaht at scripps.edu><mailto:hannaht at scripps.edu>
To: "3dem at ncmir.ucsd.edu"<mailto:3dem at ncmir.ucsd.edu> <3dem at ncmir.ucsd.edu><mailto:3dem at ncmir.ucsd.edu>
Subject: [3dem] DC EMI cancellation
Message-ID:
<PH0PR11MB74707CAF6AE2A0BE4FA9B80FAC2FA at PH0PR11MB7470.namprd11.prod.outlook.com><mailto:PH0PR11MB74707CAF6AE2A0BE4FA9B80FAC2FA at PH0PR11MB7470.namprd11.prod.outlook.com>
Content-Type: text/plain; charset="us-ascii"
Dear Colleagues,
Has anyone had to correct for DC EMI after a site survey? Are there ways to reduce or block occasional near DC peaks other than installing an entire cancellation sytem?
Thank you,
Hannah L. Turner
Lab Manager, Andrew Ward Lab
Scripps Research Institute, La Jolla CA, USA
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=http*3A*2F*2Fmail.ncmir.ucsd.edu*2Fpipermail*2F3dem*2Fattachments*2F20230705*2F6ffa512e*2Fattachment-0001.html&data=05*7C01*7Cchunyan.wang.cw785*40yale.edu*7C91bc1a7b767a4e026de808db7df95db7*7Cdd8cbebb21394df8b4114e3e87abeb5c*7C0*7C0*7C638242282236502146*7CUnknown*7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0*3D*7C3000*7C*7C*7C&sdata=NmvV9awdVn63xO6jzbyOoxN95G8w7hTm95F7GI6CtCg*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!DWcQkg0Ty4cqq9C59GJZt1mAPqlCBYBOcOpahwXVeCJqSDle1NQC0O0UXr0ojcgeckgwRADuGyu_zIMjgxC6vZjxIA$ <http://mail.ncmir.ucsd.edu/pipermail/3dem/attachments/20230705/6ffa512e/attachment-0001.html>>
------------------------------
Message: 3
Date: Wed, 5 Jul 2023 10:30:41 -0400
From: Matthew Chang <changmlw at umich.edu><mailto:changmlw at umich.edu>
To: 3dem at ncmir.ucsd.edu<mailto:3dem at ncmir.ucsd.edu>
Subject: [3dem] Going from Amira filament picks to subtomogram
averaging
Message-ID:
<CAMZ7i9R9wvrpXw7dgU-65e0mzFnoxOJDC0p3Kx39HAdKPBB2nw at mail.gmail.com><mailto:CAMZ7i9R9wvrpXw7dgU-65e0mzFnoxOJDC0p3Kx39HAdKPBB2nw at mail.gmail.com>
Content-Type: text/plain; charset="utf-8"
Hello all,
I segmented filaments in Amira manually. I want to use that segmentation to
extract coordinates with tilt and psi angle priors to do subtomogram
reconstruction in Warp and subtomogram averaging in relion (using helical
processing). Does anyone know if this is possible?
Any help would be appreciated!
Thank you,
Matthew
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=http*3A*2F*2Fmail.ncmir.ucsd.edu*2Fpipermail*2F3dem*2Fattachments*2F20230705*2Fc17dd940*2Fattachment-0001.html&data=05*7C01*7Cchunyan.wang.cw785*40yale.edu*7C91bc1a7b767a4e026de808db7df95db7*7Cdd8cbebb21394df8b4114e3e87abeb5c*7C0*7C0*7C638242282236502146*7CUnknown*7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0*3D*7C3000*7C*7C*7C&sdata=WPqXTqSiNmN*2F4i5uJ9MY9iiyHFcqgWsj0ptAZfTv9bE*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSU!!Mih3wA!DWcQkg0Ty4cqq9C59GJZt1mAPqlCBYBOcOpahwXVeCJqSDle1NQC0O0UXr0ojcgeckgwRADuGyu_zIMjgxAgfbK8fg$ <http://mail.ncmir.ucsd.edu/pipermail/3dem/attachments/20230705/c17dd940/attachment-0001.html>>
------------------------------
Message: 4
Date: Thu, 6 Jul 2023 08:31:52 +0200
From: Andrzej ?ak <andrzej.zak at pwr.edu.pl><mailto:andrzej.zak at pwr.edu.pl>
To: Hannah Turner <hannaht at scripps.edu><mailto:hannaht at scripps.edu>
Cc: 3dem at ncmir.ucsd.edu<mailto:3dem at ncmir.ucsd.edu>
Subject: Re: [3dem] DC EMI cancellation
Message-ID:
<CAJqBd=eaYzhBPKFG9Hnx1=6FTqGjUMim1yXa4uEyQ350fLXDqQ at mail.gmail.com><mailto:CAJqBd=eaYzhBPKFG9Hnx1=6FTqGjUMim1yXa4uEyQ350fLXDqQ at mail.gmail.com>
Content-Type: text/plain; charset="utf-8"
Dear Hannah,
I had such situation near the road, probably city start using other/more
trams. The main reasons of DC are trams, trains, metros, elevators, etc. If
pre measurement was taken during holidays, during the normal work day could
be different (and getting worse in time, if anyone around mount some fancy
equipment or you have more buildings and stuff around). As far as I know
field cancelation like Spicer is the only way to go for DC. And nowadays
it's good practice to mount it for any new TEM, just in case for the
future.
Best wishes
Andrzej
Andrzej ?ak, PhD Eng
Wroclaw University of Science and Technology,
tel. (+48)883314999
e-mail: andrzej.zak at pwr.edu.pl<mailto:andrzej.zak at pwr.edu.pl>
?r., 5 lip 2023, 22:06 u?ytkownik Hannah Turner <hannaht at scripps.edu><mailto:hannaht at scripps.edu>
napisa?:
> Dear Colleagues,
>
>
>
> Has anyone had to correct for DC EMI after a site survey? Are there ways
> to reduce or block occasional near DC peaks other than installing an entire
> cancellation sytem?
>
>
>
> Thank you,
>
> Hannah L. Turner
>
> Lab Manager, Andrew Ward Lab
>
> Scripps Research Institute, La Jolla CA, USA
> _______________________________________________
> 3dem mailing list
> 3dem at ncmir.ucsd.edu<mailto:3dem at ncmir.ucsd.edu>
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Fmail.ncmir.ucsd.edu*2Fmailman*2Flistinfo*2F3dem&data=05*7C01*7Cchunyan.wang.cw785*40yale.edu*7C91bc1a7b767a4e026de808db7df95db7*7Cdd8cbebb21394df8b4114e3e87abeb5c*7C0*7C0*7C638242282236502146*7CUnknown*7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0*3D*7C3000*7C*7C*7C&sdata=aCQ*2F43X3rSG0Ud493IAAAvHMKzpB*2Fg6*2BOXadE4gB6U4*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!DWcQkg0Ty4cqq9C59GJZt1mAPqlCBYBOcOpahwXVeCJqSDle1NQC0O0UXr0ojcgeckgwRADuGyu_zIMjgxCR4oLwpQ$ <https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem>
>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=http*3A*2F*2Fmail.ncmir.ucsd.edu*2Fpipermail*2F3dem*2Fattachments*2F20230706*2F4e37b031*2Fattachment-0001.html&data=05*7C01*7Cchunyan.wang.cw785*40yale.edu*7C91bc1a7b767a4e026de808db7df95db7*7Cdd8cbebb21394df8b4114e3e87abeb5c*7C0*7C0*7C638242282236502146*7CUnknown*7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0*3D*7C3000*7C*7C*7C&sdata=hCYiszjJSFS87FyUcWhg0jk6Yhkz7w0UmaaKhytsw9s*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!DWcQkg0Ty4cqq9C59GJZt1mAPqlCBYBOcOpahwXVeCJqSDle1NQC0O0UXr0ojcgeckgwRADuGyu_zIMjgxDflJlbLQ$ <http://mail.ncmir.ucsd.edu/pipermail/3dem/attachments/20230706/4e37b031/attachment-0001.html>>
------------------------------
Message: 5
Date: Thu, 6 Jul 2023 06:57:27 +0000
From: Kikuti Carlos <Carlos.Kikuti at curie.fr><mailto:Carlos.Kikuti at curie.fr>
To: "CCPEM at jiscmail.ac.uk"<mailto:CCPEM at jiscmail.ac.uk> <CCPEM at jiscmail.ac.uk><mailto:CCPEM at jiscmail.ac.uk>,
"3dem at ncmir.ucsd.edu"<mailto:3dem at ncmir.ucsd.edu> <3dem at ncmir.ucsd.edu><mailto:3dem at ncmir.ucsd.edu>
Subject: [3dem] Post-doc in Paris
Message-ID: <20DF0C54-FF62-45A4-8859-50304D20FEEC at curie.fr><mailto:20DF0C54-FF62-45A4-8859-50304D20FEEC at curie.fr>
Content-Type: text/plain; charset="us-ascii"
The Structural Motility team, directed by Anne Houdusse at the Curie Institute (Paris Center), is searching for a post-doctoral fellow.
This dynamic laboratory is looking for an expert in sample preparation and structural determination via cryoEM studies. We use a multi-disciplinary approach combining structural biology (x-ray and cryoEM), chemical biology and cell biology to innovate in the investigation and understanding of molecular mechanisms involving different motors in cells. Our studies concern the very basics of force production, as well as the different cellular roles of the motors, their regulation, and how they can be targeted for developing novel chemical biology tools and therapies against human diseases.
We are looking to recruit and train a dynamic and motivated candidate who is eager to gain independence and take the lead on one of our exciting scientific projects. Experience in sample preparation for cryoEM is essential. Experience in data processing, X-ray crystallography, molecular dynamics or cell biology is a plus.
If you are interested, please send a CV and a statement of research interests and goals as well as three letters of recommendation from your previous employers.
Contact : Anne Houdusse (anne.houdusse at curie.fr<mailto:anne.houdusse at curie.fr><mailto:anne.houdusse at curie.fr><mailto:anne.houdusse at curie.fr>)
Website : Structural Motility Team https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Finstitut-curie.org*2Fteam*2Fhoudusse__*3B!!Mih3wA!Hf9w-ttvBHK5gtdM-qUvvGg3ZI5VUJDI044eyurC9Z6mlRfTyZTyqlY5SM8y-SYMoCJhChwWRG9fjFXUyOsz0SMjS58*24&data=05*7C01*7Cchunyan.wang.cw785*40yale.edu*7C91bc1a7b767a4e026de808db7df95db7*7Cdd8cbebb21394df8b4114e3e87abeb5c*7C0*7C0*7C638242282236502146*7CUnknown*7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0*3D*7C3000*7C*7C*7C&sdata=pOSqez*2Ff*2Fo4aY17KRd9y4*2Fh9m8ccFuYW1a*2F8m1C9NKI*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSU!!Mih3wA!DWcQkg0Ty4cqq9C59GJZt1mAPqlCBYBOcOpahwXVeCJqSDle1NQC0O0UXr0ojcgeckgwRADuGyu_zIMjgxDhiiTTIg$ <https://urldefense.com/v3/__https://institut-curie.org/team/houdusse__;!!Mih3wA!Hf9w-ttvBHK5gtdM-qUvvGg3ZI5VUJDI044eyurC9Z6mlRfTyZTyqlY5SM8y-SYMoCJhChwWRG9fjFXUyOsz0SMjS58$>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=http*3A*2F*2Fmail.ncmir.ucsd.edu*2Fpipermail*2F3dem*2Fattachments*2F20230706*2F4cf1f93b*2Fattachment-0001.html&data=05*7C01*7Cchunyan.wang.cw785*40yale.edu*7C91bc1a7b767a4e026de808db7df95db7*7Cdd8cbebb21394df8b4114e3e87abeb5c*7C0*7C0*7C638242282236502146*7CUnknown*7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0*3D*7C3000*7C*7C*7C&sdata=aBRcTvhNvifPBfP9RPYyURTd38pCoOqedKw2Se2MQOM*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!DWcQkg0Ty4cqq9C59GJZt1mAPqlCBYBOcOpahwXVeCJqSDle1NQC0O0UXr0ojcgeckgwRADuGyu_zIMjgxA7vzSl2w$ <http://mail.ncmir.ucsd.edu/pipermail/3dem/attachments/20230706/4cf1f93b/attachment-0001.html>>
------------------------------
Message: 6
Date: Thu, 6 Jul 2023 08:16:12 +0000
From: Stahlberg Henning <henning.stahlberg at epfl.ch><mailto:henning.stahlberg at epfl.ch>
To: Andrzej ?ak <andrzej.zak at pwr.edu.pl><mailto:andrzej.zak at pwr.edu.pl>, Hannah Turner
<hannaht at scripps.edu><mailto:hannaht at scripps.edu>
Cc: "3dem at ncmir.ucsd.edu"<mailto:3dem at ncmir.ucsd.edu> <3dem at ncmir.ucsd.edu><mailto:3dem at ncmir.ucsd.edu>
Subject: Re: [3dem] DC EMI cancellation
Message-ID: <CD9B987F-B840-44E3-92F3-775FC6A88E9F at epfl.ch><mailto:CD9B987F-B840-44E3-92F3-775FC6A88E9F at epfl.ch>
Content-Type: text/plain; charset="utf-8"
Hi,
We had strong EM fields from a nearby metro on one side of the building and a DC cable running hundreds of amperes DC current close to the intended EM site on the other side of the building. DC field variations at the intended EM location were stronger than 4000nT, while you wish to keep that below 50nT. DC fields decay linearly with the distance from the source cable, while AC fields decay with distance squared from the source cable. So, in a certain distance, DC fields are the remaining problem.
Fields can be shielded passively with annealed mu-metal, which is very expensive and in addition requires to install a room in a room, which swallows 20cm from each wall. After that, fields inside the mu-metal cage are homogeneous and easier to shield passively. But the price is very high.
Fields can also be shielded actively with Helmholtz coils in X, Y and Z, providing a counter-field that neutralizes the field at least for the one location in the room where the system's field sensor is. That works well for a TEM without energy filter, if the field sensor of that active cancellation system is close to the sample stage, where the sensitivity is the highest. If in addition you also need a low field in a post-column filter of that TEM, then an active two-sensor / two-coil-pair system might be an option, or you need to go for the much more expensive passive shield solution. The active cancellation system compensates fields from 0Hz up to 1000Hz for one location in the room. If fields are inhomogeneous, e.g., from a nearby cable, then active cancellation would correct at one spot in the room but make matters worse everywhere else in the room.
Working closely with the excellent Systron AG in Zurich, Switzerland, we decided to move the instruments as far away from the source as possible (about 12 m in this case), and then install only the active cancellation systems MR-3 by Stefan Mayer in each EM room. This was 40x cheaper than passive cancellation and didn't consume any space. It worked well, all 300kV instruments reach 1.2? or better on apoferritin, also the one with postcolumn filter, even though the latter will benefit from the better location into which we will move the instruments soon.
Best wishes,
Henning.
Henning Stahlberg
Laboratory of?Biological Electron Microscopy
Institute of?Physics, School of?Basic Sciences, EPFL, and
Dep. of Fund. Microbiology,?Faculty of Biology and Medicine, UNIL,?
Cubotron, BSP421, 1015 Lausanne,?Switzerland
https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Flbem.ch*2F__*3B!!Mih3wA!BQM-ZCoa__DkhhGjRvCxDmh1rDex6QX_NhUrxP974UrwJe-_53jEEybrH22of4dKS3sj6qDLdO09rqpYamJl0poEFLb_a5g*24&data=05*7C01*7Cchunyan.wang.cw785*40yale.edu*7C91bc1a7b767a4e026de808db7df95db7*7Cdd8cbebb21394df8b4114e3e87abeb5c*7C0*7C0*7C638242282236502146*7CUnknown*7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0*3D*7C3000*7C*7C*7C&sdata=T3fJvVtJiNkIPdwmUEKbee5MqeYcIRCeFIwDEsLfAqQ*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUl!!Mih3wA!DWcQkg0Ty4cqq9C59GJZt1mAPqlCBYBOcOpahwXVeCJqSDle1NQC0O0UXr0ojcgeckgwRADuGyu_zIMjgxARwrQacg$ <https://urldefense.com/v3/__https://lbem.ch/__;!!Mih3wA!BQM-ZCoa__DkhhGjRvCxDmh1rDex6QX_NhUrxP974UrwJe-_53jEEybrH22of4dKS3sj6qDLdO09rqpYamJl0poEFLb_a5g$> , +41 21 693?45 07
From: 3dem <3dem-bounces at ncmir.ucsd.edu><mailto:3dem-bounces at ncmir.ucsd.edu> on behalf of Andrzej ?ak <andrzej.zak at pwr.edu.pl><mailto:andrzej.zak at pwr.edu.pl>
Date: Thursday, 6 July 2023 at 08:32
To: Hannah Turner <hannaht at scripps.edu><mailto:hannaht at scripps.edu>
Cc: 3dem <3dem at ncmir.ucsd.edu><mailto:3dem at ncmir.ucsd.edu>
Subject: Re: [3dem] DC EMI cancellation
Dear Hannah,?
I had such situation near the road, probably city start using other/more trams. The main reasons of DC are trams, trains, metros, elevators, etc. If pre measurement was taken during holidays, during the normal work day could be different (and getting worse in time, if anyone around mount some fancy equipment or you have more buildings and stuff around). As far as I know field cancelation like Spicer is the only way to go for DC. And nowadays it's good practice to mount it for any new TEM, just in case for the future.?
Best wishes?
Andrzej?
Andrzej ?ak, PhD Eng
Wroclaw University of Science and Technology,
tel. (+48)883314999
e-mail: mailto:andrzej.zak at pwr.edu.pl
?r., 5 lip 2023, 22:06 u?ytkownik Hannah Turner <mailto:hannaht at scripps.edu> napisa?:
Dear Colleagues,
?
Has anyone had to correct for DC EMI after a site survey? Are there ways to reduce or block occasional near DC peaks other than installing an entire cancellation sytem?
?
Thank you,
Hannah L. Turner
Lab Manager, Andrew Ward Lab
Scripps Research Institute, La Jolla CA, USA
_______________________________________________
3dem mailing list
mailto:3dem at ncmir.ucsd.edu
https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Fmail.ncmir.ucsd.edu*2Fmailman*2Flistinfo*2F3dem&data=05*7C01*7Cchunyan.wang.cw785*40yale.edu*7C91bc1a7b767a4e026de808db7df95db7*7Cdd8cbebb21394df8b4114e3e87abeb5c*7C0*7C0*7C638242282236502146*7CUnknown*7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0*3D*7C3000*7C*7C*7C&sdata=aCQ*2F43X3rSG0Ud493IAAAvHMKzpB*2Fg6*2BOXadE4gB6U4*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!DWcQkg0Ty4cqq9C59GJZt1mAPqlCBYBOcOpahwXVeCJqSDle1NQC0O0UXr0ojcgeckgwRADuGyu_zIMjgxCR4oLwpQ$ <https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.png
Type: image/png
Size: 17052 bytes
Desc: image001.png
URL: <https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=http*3A*2F*2Fmail.ncmir.ucsd.edu*2Fpipermail*2F3dem*2Fattachments*2F20230706*2Fbb02eaec*2Fattachment.png&data=05*7C01*7Cchunyan.wang.cw785*40yale.edu*7C91bc1a7b767a4e026de808db7df95db7*7Cdd8cbebb21394df8b4114e3e87abeb5c*7C0*7C0*7C638242282236502146*7CUnknown*7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0*3D*7C3000*7C*7C*7C&sdata=obsl7Enj5tdI6x9kt81v4ZiFnb2fLAKCSloxPaiw1tI*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!DWcQkg0Ty4cqq9C59GJZt1mAPqlCBYBOcOpahwXVeCJqSDle1NQC0O0UXr0ojcgeckgwRADuGyu_zIMjgxDsQiIGfw$ <http://mail.ncmir.ucsd.edu/pipermail/3dem/attachments/20230706/bb02eaec/attachment.png>>
------------------------------
Subject: Digest Footer
_______________________________________________
3dem mailing list
3dem at ncmir.ucsd.edu<mailto:3dem at ncmir.ucsd.edu>
https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Fmail.ncmir.ucsd.edu*2Fmailman*2Flistinfo*2F3dem&data=05*7C01*7Cchunyan.wang.cw785*40yale.edu*7C91bc1a7b767a4e026de808db7df95db7*7Cdd8cbebb21394df8b4114e3e87abeb5c*7C0*7C0*7C638242282236502146*7CUnknown*7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0*3D*7C3000*7C*7C*7C&sdata=aCQ*2F43X3rSG0Ud493IAAAvHMKzpB*2Fg6*2BOXadE4gB6U4*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!DWcQkg0Ty4cqq9C59GJZt1mAPqlCBYBOcOpahwXVeCJqSDle1NQC0O0UXr0ojcgeckgwRADuGyu_zIMjgxCR4oLwpQ$ <https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem>
------------------------------
End of 3dem Digest, Vol 191, Issue 7
************************************
_______________________________________________
3dem mailing list
3dem at ncmir.ucsd.edu<mailto:3dem at ncmir.ucsd.edu>
https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://mail.ncmir.ucsd.edu/pipermail/3dem/attachments/20230706/f857c41b/attachment-0001.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: ColdTrap01.jpg
Type: image/jpeg
Size: 214148 bytes
Desc: ColdTrap01.jpg
URL: <http://mail.ncmir.ucsd.edu/pipermail/3dem/attachments/20230706/f857c41b/attachment-0004.jpg>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: ColdTrap02.jpg
Type: image/jpeg
Size: 218363 bytes
Desc: ColdTrap02.jpg
URL: <http://mail.ncmir.ucsd.edu/pipermail/3dem/attachments/20230706/f857c41b/attachment-0005.jpg>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: ColdTrap03.jpg
Type: image/jpeg
Size: 172682 bytes
Desc: ColdTrap03.jpg
URL: <http://mail.ncmir.ucsd.edu/pipermail/3dem/attachments/20230706/f857c41b/attachment-0006.jpg>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: ColdTrap04.jpg
Type: image/jpeg
Size: 390264 bytes
Desc: ColdTrap04.jpg
URL: <http://mail.ncmir.ucsd.edu/pipermail/3dem/attachments/20230706/f857c41b/attachment-0007.jpg>
More information about the 3dem
mailing list