Craig,
thanks!  that gives me a good place to start…
i’ll let you know what we see here when pushing the FPS up above 40.
Garry


On Feb 14, 2019, at 12:57 PM, Craig Yoshioka <yoshiokc at ohsu.edu<mailto:yoshiokc at ohsu.edu>> wrote:

Hi Gary,

We haven’t exhaustively tested yet, but we have been targeting around 15-20eps on the K3 and have been getting nice results.  This general range was based on the 400 -> 1500 read rate increase.

At very high-magnifications 15eps is quite a high flux, so sometimes we will go below 10eps just so that we don’t have to push very high FPS (40+) with < 1 sec exposures.  We have seen some mild pattern noise occasionally creep into images when trying to save at high frame rates.  Not sure yet if this is just our K3, or if it also seen by others.  So far it doesn’t seem to be a problem in practice, but feedback from others if they have seen this would be appreciated.

Cheers,
-Craig


On Feb 14, 2019, at 11:45 AM, Garry P Morgan <Garry.Morgan at Colorado.EDU<mailto:Garry.Morgan at Colorado.EDU>> wrote:

thanks everyone.

On Feb 14, 2019, at 11:53 AM, Reyes, Francis <francis.reyes at thermofisher.com<mailto:francis.reyes at thermofisher.com>> wrote:

Yes it is advisable to work in nanoprobe. At parallel illumination and a 50 um C2 you should have around a 1.8um illuminated area.


Francis Reyes, Ph.D
Engineer III, Field Applications
Materials & Structural Analysis

Thermo Fisher Scientific
5350 NE Dawson Creek Drive<x-apple-data-detectors://37>
Hillsboro, OR 97124<x-apple-data-detectors://37>
Phone - +1 (720) 322-6150<tel:+1%20(720)%20322-6150>
francis.reyes at thermofisher.com<mailto:francis.reyes at thermofisher.com>
thermofisher.com<http://thermofisher.com/>

On Feb 14, 2019, at 12:36 PM, Garry P Morgan <Garry.Morgan at colorado.edu<mailto:Garry.Morgan at colorado.edu>> wrote:

CAUTION: This email originated from outside of the organization. Do not click links or open attachments unless you recognize the sender and know the content is safe.


Hello all,

i’m looking for some info on the optimal dose rate for acquiring K3 movie stacks.  i would appreciate it if any of you could send me your optimum acquisition parameters for imaging cry-samples for SPA…ie, dose rate, best frame rate/number of frames, etc.

i’m also trying to figure out how to setup my low dose so that we have parallel illumination for our record images (on an FEI Tecnai F20). the problem i see is that in microprobe the beam is spread much too wide to image/expose a single hole at a time at our record mag. is it advisable to work in nano probe to get a smaller beam that is parallel? any advice on this would be appreciated.

thanks,
Garry



________________________________
Garry Morgan
Director  —  Boulder EM Services —  Department of MCD Biology
Campus Box 347  —  University of Colorado  —  Boulder, CO 80309
phone:  303-492-8402

_______________________________________________
3dem mailing list
3dem at ncmir.ucsd.edu<mailto:3dem at ncmir.ucsd.edu>
https://urldefense.proofpoint.com/v2/url?u=https-3A__mail.ncmir.ucsd.edu_mailman_listinfo_3dem&d=DwIGaQ&c=q6k2DsTcEGCcCb_WtVSz6hhIl8hvYssy7sH8ZwfbbKU&r=sQUVJjPZfltp7Dvz0Uv6lCoV0pjSGOuvcvkq06OG19I&m=nO-BGOChroWTiPhZriOs8GEf78gDKCbKAcfxxONJ48w&s=snNp1N1J6yQFOr8-P30FANtvBYCx7SczYBLjyYvWjP8&e=

_______________________________________________
3dem mailing list
3dem at ncmir.ucsd.edu<mailto:3dem at ncmir.ucsd.edu>
https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem


-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://mail.ncmir.ucsd.edu/pipermail/3dem/attachments/20190214/379bb56d/attachment.html>


More information about the 3dem mailing list