[3dem] Cryoplunge 3 question
Matthias Wolf
matthias.wolf at oist.jp
Thu Feb 7 05:02:51 PST 2019
Hi David,
we also have a CP3 and I have never seen this problem.
I would check that there is sufficient N2 pressure. The mechanism is pneumatic. At low pressure it can get stuck or not move all the way. You could try to increase the gas pressure at your N2 cylinder regulator and check for any leaks at the fittings.
Best regards,
Matthias
________________________________
From: 3dem <3dem-bounces at ncmir.ucsd.edu> on behalf of Morgan, David Gene <dagmorga at indiana.edu>
Sent: Thursday, February 7, 2019 7:01 AM
To: 3dem at ncmir.ucsd.edu
Subject: [3dem] Cryoplunge 3 question
Hi,
We are currently demo'ing a Gatan Cryoplunge 3 (CP3) freezing robot. From time to time (and randomly as far as we can tell), there is a significant delay between the time the start button is pushed and when the blotters actually close. Since the countdown for blotting time starts when the button is pushed, the delay causes the actual blot time for that particular grid to be too short. Does anyone else who has a CP3 see this issue, and if so, is there anything that can be done about it? Thanks in advance.
--
politics is more difficult than physics.
A. Einstein
David Gene Morgan
Electron Microscopy Center
047D Simon Hall
IU Bloomington
812 856 1457 (office)
812 856 3221 (3200)
http://iubemcenter.indiana.edu
_______________________________________________
3dem mailing list
3dem at ncmir.ucsd.edu
https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://mail.ncmir.ucsd.edu/pipermail/3dem/attachments/20190207/a8441516/attachment-0001.html>
More information about the 3dem
mailing list