Hi Michael,
It's made to load grids in gas, but many overfill to work submerged, then it definitely breaks. Heating too much also is a bad idea. No issues here with the seven year old station, the two new ones delivered last year are also (still...) fine.
Best,
Wim
> On Sep 6, 2018, at 17:59, Michael Braunfeld <braun at msg.ucsf.edu> wrote:
>
> Hi All,
>
> Do other facilities have any issues with the loading stations for the Krios/Arctica? We just experienced our second failure in the last month. The LN2 is leaking out at a pretty rapid rate from cracks at the base of the stations. These are not being abused in any way (as far as I know), and the worst part is the price quote from Thermo.
>
>
> Thanks,
>
> Michael Braunfeld
>
> HHMI/UCSF
>
> <IMG_9779.JPG>