We had problems with a batch of C-flats that we got late last year. The
carbon broke apart if we used the normal glow discharge process that we
used for earlier batches and we had to greatly reduce the time that we
used to get most of the squares intact after plunging. Our carbon on
these grids also disintegrated under a focused beam. We also observed a
greater incidence of beam induced charging on these samples. You might
want to contact the people at Protochips. They seem to be aware of this
'thinning' problem with the carbon and they could probably tell you if
your batch might be affected.
Cheers,
Jamie
Thorsten Mielke wrote:
> Dear all,
>
> we are trying to vitrify some small proteins on c-flat grids (type
> CF-2/4-4C) using a Vitrobot from FEI. However, after plunge-freezing,
> the carbon film is heavily destroyed. Additionally, we find only a few
> good holes on the remaining intact meshes. The carbon film also seems to
> be extremely fragile and easily disrupts upon focusing (300 kV, 155kx
> Polara). The grids seem to be OK (with or without glow-discharching)
> until plunge-freezing. We varied the blotting offset at the Vitrobot to
> play with the blotting force, no success. Did you observe similar
> problems with C-flats? Any idea how to prevent the carbon film from
> breaking?
>
> Thanks
>
> Thorsten
>
>
>
>
--
--------------------