[3dem] RE: EM sample in high salt
Peijun Zhang
pez7+ at pitt.edu
Wed Jul 9 10:08:12 PDT 2008
Thanks to all that replied. You might find this useful.
The question was: How to make good cryoEM specimen for the sample only
stable in high salt (1M Nacl)? The following is a summary of the replies.
1) use GraFix method is: Kastner B et al. GraFix: sample preparation
for single-particle electron cryomicroscopy. Nat Methods. 2008
Jan;5(1):53-5. Epub 2007 Dec 23.
2) cross-linking with low glutaraldehyde concentrations.
3) freeze the sample in high salt. Nature (2001), vol 409 1047 (note,
S/N was actually poor).
4) using the cryo-negative staining protocol at (~0.84M Ammonium
Molybdate) developed by Sacha De Carlo et al.
5) dilution on a holey grid: put your sample onto one side of a holey
grid and put a large droplet of buffer at the desired salt concentration (or
even a bit lower) on the opposite side of the grid. Then blot from the back
side (the side with the large buffer droplet)
Best,
Peijun
_____
From: Peijun Zhang [mailto:pez7+ at pitt.edu]
Sent: Monday, July 07, 2008 6:16 PM
To: '3dem at ncmir.ucsd.edu'
Subject: EM sample in high salt
Hello all,
I have an EM specimen which only stabilizes in high salt buffer (1M NaCl).
I would like to carry out cryoEM imaging at lower salt conditions. Does
anyone have a good protocol to reduce salt to 100mM or less without
compromising the specimen?
Thanks so much,
Best,
Peijun
-------------- next part --------------
An HTML attachment was scrubbed...
URL: https://mail.ncmir.ucsd.edu/mailman/private/3dem/attachments/20080709/e1a7aebc/attachment.html
More information about the 3dem
mailing list