>Can someone remind me of typical operating conditions for the electron
>guns for shadowing at 45 degrees with PtC and then C coating for an old
>style Balzers machine (BAF series). I would like to know the voltage and
>current and approximate time settings please!
According to my old notebook (when I was in Nobutaka Hirokawa's lab),
wait until the vacuum become 10-6 to 10e-7 torr
cool down the sample to maximum (0-0)
fracture (this process can be difficult..) do not fracture too deep.
etching at -95C (dial 5-30) for 6min30sec (after the sample reached -95C)
evaporate platinum (filament 6.5, high voltage 5 (1750V) )
at this point, surface of the sample is covered by the knife.
filament to 1
stage rotary on ( rotation speed is 8 )
after uncovering the sample by moving knife, evaporate platinum up to
1.5nm and switch of the filament.
evaporate carbon for 5sec
take out sample
Some of the setting depends on the individual machine, so play around the
value described here.
Good luck !
--
-------------------------------------------------------