<div dir="ltr"><div dir="ltr">Hi Tim, we use Elsa 698 with the Gatan 1905 controller and 655 pumping station. The bakeout temperature is kept at 100C, never exceeding it for more than a few degrees. We usually perform overnight bakeout once a week, but we pump out the holder dewar after each cryo-day. I think that it is ok for the holder to perform bakeout more often in case of extensive sample exchange. You definitely need to do the holder bakeout if:<div>1) The holder fails to reach the -170 C temperature within reasonable time. In our case ~45-50 min inside the pumping station, but may be different for your holder.<div>2) The holder cannot keep the cold temperature during operation </div><div>3) The frost or condensate forms on the holder dewar </div><div>4) The holder was not used for a couple of weeks. </div></div><div>In case the holder was not used for a longer time or in case the overnight bakeout does not solve the temperature problem, the longer bakeout times should be used, like 48h or 72h.</div></div><br><div class="gmail_quote"><div dir="ltr" class="gmail_attr">пн, 29 авг. 2022 г. в 15:52, Tim Gruene <<a href="mailto:tim.gruene@univie.ac.at" target="_blank">tim.gruene@univie.ac.at</a>>:<br></div><blockquote class="gmail_quote" style="margin:0px 0px 0px 0.8ex;border-left:1px solid rgb(204,204,204);padding-left:1ex">Dear all,<br>
<br>
we recently received our Gatan Elsa 698 holder. I would like to gather<br>
some recommendations for daily use and bake-out procedure. <br>
- what temperature do you use for bake-out?<br>
- how long do you bake-out for?<br>
- are there special maintenance procedures?<br>
- what would be signs of deterioriation, abd after what period would<br>
  they occur?<br>
<br>
We use the holder both for room-temperature and for cryo-measurements.<br>
<br>
The manual refers to the Gatan pumping station, but ours was delivered<br>
with a Pfeiffer pumping station ;-)<br>
<br>
Thanks a lot for your time!<br>
<br>
Best,<br>
Tim<br>
<br>
-- <br>
--<br>
Tim Gruene<br>
Head of the Centre for X-ray Structure Analysis<br>
Faculty of Chemistry<br>
University of Vienna<br>
<br>
Phone: +43-1-4277-70202<br>
<br>
GPG Key ID = A46BEE1A<br>
_______________________________________________<br>
3dem mailing list<br>
<a href="mailto:3dem@ncmir.ucsd.edu" target="_blank">3dem@ncmir.ucsd.edu</a><br>
<a href="https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem" rel="noreferrer" target="_blank">https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem</a><br>
</blockquote></div><div><br></div><div dir="ltr"><div dir="ltr">--<br></div></div><div dir="ltr" class="gmail_signature" data-smartmail="gmail_signature"><div dir="ltr"><p class="MsoNormal" style="margin-bottom:10.0pt;line-height:115%">Best regards,<br>
Andrey Moiseenko</p>

<span style="font-size:11.0pt;font-family:"Calibri",sans-serif">Electron Microscopy Lab @ Biology Department<br>
Moscow State University<br>
<a href="mailto:postmoiseenko@gmail.com" target="_blank"><span style="color:blue">postmoiseenko@gmail.com</span></a></span><br></div></div></div>