<div dir="ltr">Good afternoon,<div><br></div><div>Some months ago we purchased a Gatan 910 multi-specimen cryo holder (three positions). It initially worked wonderfully, but after some time we started to observe that some of the grids would melt inside the microscope. Sometimes they would be melted from the very beginning and sometimes it would happen randomly during the EM session.</div><div><br></div><div>After some troubleshooting we have observed that, over time, the splitrings tend to <span style="color:rgb(0,0,0);font-family:Helvetica;font-size:12px">get too constricted</span>, almost closing the C-shaped gap (pic attached). The splitrings then sit too loose on the holder (they can even rotate in place when they are clipped), and they do not make proper contact with the grid, causing the temperature loss.</div><div><br></div><div>We now regularly open up the splitrings a little (with a pipette tip) and that seems to fix the problem. </div><div><br></div><div>We hope this might be helpful (and save some frustration) to other Gatan 910 users.</div><div><br></div><div>Best regards,</div><div>Ernesto.</div><div><br></div><div><br></div><div>--<br>Ernesto Arias Palomo, PhD<br>Group Leader<br>Center for Biological Research Margarita Salas (CSIC)<br>Ramiro de Maeztu, 9.<br>28040. Madrid. Spain<br><a href="mailto:earias@cib.csic.es">earias@cib.csic.es</a><br>+34 91 837 31 12 ext 4446<br></div><div><br></div><div><br></div></div>