We had leopard ice on our 914 holder all the time, and we had to send it in for repair in order for the problem to be solved. The shutter must close all the way as well. <br>
<br>
<div id="AppleMailSignature">Sent from my iPhone</div>
<div><br>
On 10 Aug 2018, at 16:53, Wim Hagen <<a href="mailto:hagen@embl.de">hagen@embl.de</a>> wrote:<br>
<br>
</div>
<blockquote type="cite">
<div>
<div></div>
<div>Dear Teige,</div>
<div><br>
</div>
<div>Leopard print is typically caused by temperature problems, when the holder not being baked properly, or the clipping not making full contact.</div>
<div>Shutter problems would show contamination by large ice crystals picked up during holder insertion through air with the shutter not completely shielding.</div>
<div><br>
</div>
<div>Best,</div>
<div><br>
</div>
<div>Wim Hagen</div>
<div>EMBL Heidelberg</div>
<div><br>
On Aug 10, 2018, at 3:44 PM, Matthews-Palmer, Teige Rowan Seal <<a href="mailto:t.matthews-palmer14@imperial.ac.uk">t.matthews-palmer14@imperial.ac.uk</a>> wrote:<br>
<br>
</div>
<blockquote type="cite">
<div>Dear 3DEMers,
<div class=""><br class="">
</div>
<div class="">While troubleshooting some “leopard print” ice contamination on a Gatan 914 holder, we noticed that the <b class="">underside</b> of the cryoshield/shutter does not close 'all the way' (you can see the grid).</div>
<div class="">I wanted to know from other users of 914 holders, does the underside of your shutter look like this when closed, or does it slide further to completely seal away the grid?</div>
<div class=""><br class="">
</div>
<div class="">An alternative cause of the contamination is a leak through the shutter rod assembly.</div>
<div class="">Any thoughts welcome!</div>
<div class="">Thanks,</div>
<div class="">Teige</div>
<div class=""><br class="">
</div>
<div class=""><20180810_115423.jpeg></div>
<div class=""><20180810_115707.jpeg></div>
</div>
</blockquote>
<blockquote type="cite">
<div><span>_______________________________________________</span><br>
<span>3dem mailing list</span><br>
<span><a href="mailto:3dem@ncmir.ucsd.edu">3dem@ncmir.ucsd.edu</a></span><br>
<span><a href="https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem">https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem</a></span><br>
</div>
</blockquote>
</div>
</blockquote>
<blockquote type="cite">
<div><span>_______________________________________________</span><br>
<span>3dem mailing list</span><br>
<span><a href="mailto:3dem@ncmir.ucsd.edu">3dem@ncmir.ucsd.edu</a></span><br>
<span><a href="https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem">https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem</a></span><br>
</div>
</blockquote>
</body>
</html>