    <div class="moz-cite-prefix">Hi,<br>
      <br>
      if the effect shows up with one out of two cryo holders only (is
      the room temperature holder giving good FFTs?) , than most likely
      something is wrong with that holder. One reason might be that the
      clip ring does not clamp the grid properly. Do you always use the
      same clip ring with each holder? Did you check the that holder at
      room temperature? If OK at RT, something might be wrong with the
      thermal bridge between the dewar and the tip. Is the cut-off
      always in the same direction?<br>
      <br>
      Best,<br>
      <br>
      Thorsten<br>
      <br>
      <br>
      <br>
      <br>
      Am 29.12.2017 um 20:23 schrieb Garry P Morgan:<br>
    </div>
    <blockquote
      cite="mid:D54771D8-1D6A-4C65-871D-8E91B1ADFB57@colorado.edu"
      type="cite">
      <meta http-equiv="Content-Type" content="text/html;
        charset=windows-1252">
      Hi,
      <div>we’re getting consistently clipped FFT’s on our TF30…the
        attached image is from a 50-frame, summed/aligned image of the
        carbon (of a cryo-grid). the drift numbers are reasonably low,
        there is no astigmatism.  just wondering if any of you can
        provide a possible explanation of this?  it definitely looks
        like drift or some sort of movement, right?  we have 2 Gatan 626
        holders, and it seems to only be happening with one of the
        holders…has anyone encountered a problem like this associated
        with a holder? </div>
      <div>thanks,</div>
      <div>Garry</div>
      <div><br>
      </div>
      <div>
        <div style="margin: 0px;"><img alt="unknown.jpg"
            apple-inline="yes" id="0E00C0C3-2DA3-42FE-90D4-B8C0D9B51754"
            apple-width="yes" apple-height="yes"
            src="cid:part1.B56C14A6.864D59DE@molgen.mpg.de"
            height="1280" width="960"></div>
      </div>
      <br>
      <fieldset class="mimeAttachmentHeader"></fieldset>
      <br>
      <pre wrap="">_______________________________________________
3dem mailing list
<a class="moz-txt-link-abbreviated" href="mailto:3dem@ncmir.ucsd.edu">3dem@ncmir.ucsd.edu</a>
<a class="moz-txt-link-freetext" href="https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem">https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem</a>
</pre>
    </blockquote>
    <br>
    <p><br>
    </p>
    <pre class="moz-signature" cols="72">-- 
*******************************************
Dr. Thorsten Mielke
- Head Microscopy and Cryo-Electron Microscopy Group -

Max Planck Institute for Molecular Genetics       
Ihnestr. 63-73                       
D-14195 Berlin                    
                           
Tel.: +49-(0)30 8413-1644
Fax: +49-(0)30 8413-1385
E-mail: <a class="moz-txt-link-abbreviated" href="mailto:mielke@molgen.mpg.de">mielke@molgen.mpg.de</a> </pre>
  </body>
</html>