[3dem] two Talos Arctica questions

Grigory Sharov sharov.grigory at gmail.com
Tue Nov 14 13:26:13 PST 2023


Hi,

#2 is a normal behaviour. TMPs generate a lot of heat and need cooling.
Some will have a fan attached, others will have water cooling. The latter
are used by TFS (I suppose) to reduce noise and possibly vibration. So if
you shut the water to your microscope, you are supposed to switch off
turbos otherwise they'll overheat

Grigory.

On Tue, 14 Nov 2023, 21:07 Morgan, David Gene, <dagmorga at indiana.edu> wrote:

> Hi,
>
> We continue to have two annoying problems with our Arctica, and I'd like
> to know whether this is just our microscope or if other have the same
> problems...
>
>    1. The minor issue is that the instrument seems to often forget the
>    position of the objective aperture.  The UI reports that the aperture
>    status is "unknown" and multiple things will not happen when this occurs.
>    We can sometimes (but not always) reset this by telling the instrument to
>    insert the aperture and then to remove it.  Our service engineer tells us
>    that the solution is to do this insert/retract step, but I am wondering
>    whether this really means that something in our aperture hardware/software
>    is not working properly.  Any experiences that others have would be
>    appreciated.
>    2. The much more major issue involves what happens when there is an
>    interruption to the chilled water flow.  On our instrument, the optics
>    board immediately overheats and the camera chillers follow that shortly.
>    Those are minor issues that can be easily fixed regardless of how long the
>    chilled water is off.  However, if the chilled water is off for "long
>    enough," the turbo pump in the autoloader shuts off and cannot/will not
>    restart.  This in turn means that the dewars cannot refill, and that the
>    autoloader warms up.  It also means that the cassette and any grids in the
>    instrument will warm up because the cassette cannot be removed - without
>    the TMP, the vacuum system can't work properly, and a grid can neither be
>    taken off the stage, nor can the entire cassette be unloaded.  This problem
>    with the TMP seems like it is either a serious design flaw in the
>    instrument (if this is what others experience) or that something is
>    fundamentally wrong with our instrument/turbo.  And in case anyone is
>    wondering, by the time we know that the chilled water will be out for long
>    enough to worry about samples in the microscope, the turbo has quit and
>    it's impossible to salvage the grids.
>
>
> Any thoughts would be greatly appreciated.
>
> --
>     politics is more difficult than physics.
>                                              A. Einstein
>
>             David Gene Morgan
>         Electron Microscopy Center
>              047E Simon Hall
>              IU Bloomington
>           812 856 1457 (office)
>           812 856 3221 (3200)
>       https://urldefense.com/v3/__http://iubemcenter.indiana.edu__;!!Mih3wA!AOq8tfSs6vhPidpNzgRxCgkYzBe_2m7eIodlkV_fh-7AsNR_KdOPVrM6FhLKy7ArsfwXqiLWpdMwASrOTpMgz9duoWM$ 
> <https://urldefense.com/v3/__http://iubemcenter.indiana.edu__;!!Mih3wA!GO-4plo-39YzB7bNOTdnH_GWKkUAPDMP-geWXhR4hQWJphA0hMy6K-BhgYgz-bvZyIN9pMDMZnJYOLIcWDEu-mQ$>
> _______________________________________________
> 3dem mailing list
> 3dem at ncmir.ucsd.edu
> https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem
>