[3dem] anisotropic magnification distortion correction with mag_distortion_estimate/correct

Xian, Yuejiao yxian at miners.utep.edu
Fri Aug 14 13:43:37 PDT 2020


Hello, Tim,

Thank you for your quick feedback. I did compare the rings and find that the change is very minor as you pointed out. However, I find it hard to believe that the distortion is so minor because many datasets were shown had distortion above 1.5%. I am wondering if the drifting would affect the estimation, as you can see that the ring is much brighter in the horizontal direction?

I am also wondering if the mag_distortion_estimate can work with the water rings Marin mentioned in his email? The water ring appears at  3.6?, 2.2? and 1.8?. If I was right, the mag_distortion_estimate only consider the area encapsulating the gold rings (2.4?-2.0?), then I should be able to perform estimation on the 2.2? water ring?

Many thanks again for your time.

Yuejiao

________________________________
From: Tim Grant <timothy.r.grant at gmail.com>
Sent: Friday, August 14, 2020 12:53 PM
To: Xian, Yuejiao <yxian at miners.utep.edu>
Cc: 3dem at ncmir.ucsd.edu <3dem at ncmir.ucsd.edu>; Yuejiao Xian <xianyuejiao at gmail.com>
Subject: Re: [3dem] anisotropic magnification distortion correction with mag_distortion_estimate/correct

Hi Yuejiao,

0.15% is not a significant result, as it changes the rings so little, so the angle changing is likely to be a random effect. Both results are essentially telling you that there is very little to no magnification distortion.  A good way to confirm this is to overlay the rotationally averaged ring with the original ring and flick between them.  If there is appreciable distortion you should see a directional difference in the rings (it will look like they are being squashed in one direction).  In your case, as far as I can tell your rings line up well.

Thanks,

Tim

On Fri, 14 Aug 2020 at 12:50, Xian, Yuejiao <yxian at miners.utep.edu<mailto:yxian at miners.utep.edu>> wrote:

Hello, all.


I am trying to estimate the anisotropic magnification distortion from a cryo-EM dataset, however, I am not sure if I got the accurate estimation. I appreciate any comments or suggestions.  Here is what I did:


I stacked about 70 gold grating images into one mrc file, which I used as the input of the mag_distortion_estimate program developed by Timothy Grant and Nikolaus Grigorieff. Here is the output parameter:

=====================================

The following distortion parameters were found :-



Distortion Angle     = 112.6

Major Scale          = 1.001

Minor Scale          = 0.999



 Stretch only parameters would be as follows :-



Distortion Angle     = 112.6

Major Scale          = 1.002

Minor Scale          = 1.000

Corrected Pixel Size = 0.669



The Total Distortion = .15%

=========================================

 The outputted ring images are attached. The ring on the left is the combined ring before any distortion correction, the one in the middle is the rotational averaged ring, the one in the right is the ring after distortion correction.


To verify if the parameters are accurate,  I then took the three parameters ( highlighted in bold) and the same stack file as an input for mag_distortion_correct to correct the distortion. With the corrected stack file, I then estimated the distortion again, and it showed the images still have 0.15% distortion, with the following parameters:

=========================================

Distortion Angle     = 21.0

Major Scale          = 1.001

Minor Scale          = 0.999.



 Stretch only parameters would be as follows :-



Distortion Angle     = 21.0

Major Scale          = 1.002

Minor Scale          = 1.000

Corrected Pixel Size = 0.669



The Total Distortion = .15%

=========================================


Note that the values of both major and minor scale remind the same as before its corrected, but the distortion angle changed dramatically. Why could be the reason for this? If you had experience with the mag_distortion_estimate/correct, would you please let me know if there is anything that I should specifically pay attention to in order to get the accurate estimation?


Many thanks in advance.


Yuejiao Xian

Ph.D candidate,
Department of Chemistry and Biochemistry,
University of Texas at El Paso
_______________________________________________
3dem mailing list
3dem at ncmir.ucsd.edu<mailto:3dem at ncmir.ucsd.edu>
https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem<https://urldefense.com/v3/__https://nam01.safelinks.protection.outlook.com/?url=https*3A*2F*2Fmail.ncmir.ucsd.edu*2Fmailman*2Flistinfo*2F3dem&data=01*7C01*7Cyxian*40miners.utep.edu*7Cbf71f12181d44e5c337508d8408bba21*7C857c21d21a1643a490cfd57f3fab9d2f*7C1&sdata=uWDSgIaT8bl1AZpaXkB1NKEMkH7Ec5tWIvJT*2Bs0MRR0*3D&reserved=0__;JSUlJSUlJSUlJSUlJSU!!Mih3wA!XjJPCk63hJqjGFCEcbF9I6VjSUZywtJQoN3Qv6cozs9Lx2nSBmNj7pGBQV2-93XsBQ$ >
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://mail.ncmir.ucsd.edu/pipermail/3dem/attachments/20200814/21eaf48d/attachment.html>


More information about the 3dem mailing list