Hi Mrs Wright,

to comment the previous answer of Mr Grassucci:

As I said I have the same blobbies, looking exactly like this, but they are in diameter roughly 30-40 nm, and I am using a vitrobot mark IV with humidity chamber set to 100%. As I also said when I am preparing grids in a row, I experience this only at the end of my grid series preparation. Thus, the first grids had the same conditons regarding the humidifier box and didn't show up such bloobies. So I dont think that it has anything to do with spray contamination of the humidifier box. I assume that it looks more like contamination by plunge-freezing through liquid ethane or kind of post-contamination when transferring it to the grid box through the liquid nitrogen atmosphere. But it is still not clear to me what is really causing such a ice formation. 

Best regards,
C. Geiss

 

----------------ursprüngliche Nachricht----------------- 
Von: "Bob Grassucci" rg2502 at columbia.edu 
An: 3dem at ncmir.ucsd.edu, "Elizabeth Wright" erwrigh at emory.edu 
Datum: Tue, 27 Sep 2011 18:56:13 -0400 
----------------------------------------------------------

>
>Hi Liz,
>Definitely not ethane. The contrast of blobs is too low. It could be coming from the 
>humidifier in the Vitrobot. Try turning off the humidifier when you are making the grids and see if 
>they disappear. I am thinking they are getting deposited from the spray while plunging.
>Bob
>
>On 9/27/2011 6:46 PM, Wright, Elizabeth R. wrote: 
>>
>>
>>
>>
>>Dear All,
>>
>> 
>>
>>For those of you patient enough to look at a larger image, please find it located at the 
>>following link:
>>
>>[http://dl.dropbox.com/u/23095990/Caulobacter_2sept2011.tif
>>
>> 
>>
>>The contamination is the small globular looking material distributed throughout 
>>the ice. There are 10 nm gold fiducials and a bacterial cell. To our eyes, when we generate 
>>the 3D reconstructions, the 'blobs' appear to be in the ice, albeit at or close to the 
>>surface. It does not look like frost or particles that have landed on the surface of a frozen 
>>specimen.
>>
>> 
>>
>>Thank you all for your help.
>>
>> 
>>
>>Sincerely,
>>
>> 
>>
>>Liz
>>
>>
>>
>> 
>>
>>
>>From: Elizabeth Wright erwrigh at emory.edu erwrigh at emory.edu 
>>Date: Tue, 27 Sep 2011 16:38:47 -0400
>>To: 3dem at ncmir.ucsd.edu 3dem at ncmir.ucsd.edu 
>>Subject: Ice contamination
>>
>> 
>>
>>
>>
>>
>>
>>Dear 3D EM friends,
>>
>> 
>>
>>We are experiencing a problem with ice contamination within the ice of our 
>>plunge-frozen samples (frozen with an FEI Vitrobot). We have purchased new hoses, regulators, 
>>ethane (99.999% pure) and liquid nitrogen tanks, attempted chloroform or methanol 
>>treatment of the grids, glow discharged the grids with different units, and still seem to be 
>>unable to remove this contaminant. It is not specimen dependent as we use different 
>>buffering systems for the samples. It is also not microscope dependent, we have seen it when 
>>using a variety of instruments.
>>
>> 
>>
>>Has anyone observed this before (image below)? Has anyone found a way to remove it? Any 
>>thoughts or suggestions would be greatly appreciated. If the image does not come through, 
>>please let me know.
>>
>> 
>>
>>Thank you all for your help.
>>
>> 
>>
>>Sincerely,
>>
>> 
>>
>>Liz
>>
>> 
>>
>> 
>>
>>
>>
>> 
>>
>>
>>----
>>
>>Elizabeth R. Wright, PhD
>>Assistant Professor
>>Emory University
>>erwrigh at emory.edu erwrigh at emory.edu 
>>Ph: (404) 727-4665
>>
>>
>>
>>
>>
>>
>>
>>__________________________________________________
>>
>>This e-mail message (including any attachments) is for the sole use of
>>the intended recipient(s) and may contain confidential and privileged
>>information. If the reader of this message is not the intended
>>recipient, you are hereby notified that any dissemination, distribution
>>or copying of this message (including any attachments) is strictly
>>prohibited.
>>
>>If you have received this message in error, please contact
>>the sender by reply e-mail message and destroy all copies of the
>>original message (including attachments).
>> 
>>
>>
>>_______________________________________________
>>3dem mailing list
>>3dem at ncmir.ucsd.edu 3dem at ncmir.ucsd.edu 
>>https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem -> 
>>http://dl.dropbox.com/u/23095990/Caulobacter_2sept2011.tif]
>>
>> 
>>
>>
>>-- 
>>*********************************
>>Howard Hughes Medical Institute
>>Columbia University Medical Center
>>Black Building Room 2-221
>>650 West 168th Street
>>New York, NY 10032-3702
>>rg2502 at columbia.edu rg2502 at columbia.edu 
>>Phone: 212-305-9520
>>Fax: 212-305-9500
>>
>>
>>__________________________________________________
>>
>>
>>_______________________________________________
>>3dem mailing list
>>3dem at ncmir.ucsd.edu 
>>https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem
>>
>>
>>
-- 

Msc. Christian Geiss
Universitaet Frankfurt FMLS 


Frangakis Gruppe 


Max-von-Laue-Strasse 15 


D - 60438 Frankfurt am Main 


Phone: +49-(0)69-79842597 


Fax: +49-(0)69-79846467 


geiss at biophysik.org 



-------------- next part --------------
An HTML attachment was scrubbed...
URL: https://mail.ncmir.ucsd.edu/mailman/private/3dem/attachments/20110928/4faf075b/attachment.html


More information about the 3dem mailing list