Kodak SO163 Dose vs OD behavior

Bob Grassucci bobg at wadsworth.org
Wed Jan 8 13:55:35 PST 2003


Hi all,
         Happy New Year.  I am reporting here the results of the 
calibration of old vs new Kodak SO163 emulsion.  At 300 kV and at 
38,500  developed 12 minutes in full strength D19 at 20 degrees C the new 
SO163 is about 10% lower density than the old emulsion.
    * At 10 e/A^2 the OD is 0.7472 with the old emulsion and 0.7104 with 
the new emulsion
    * At 20 e/A^2 the OD is 1.3146 with the old and 1.1836 with the new
    * At 30 e/A^2 the OD is 1.7595 with the old and 1.6343 with the new
    * At 40 e/A^2 the OD is 2.1453 with the Old and 1.9976 with the new
    * At 50 e/A^2 the OD is 2.4372 with the Old and 2.2934 with the new
It seems almost as if there is slightly less silver in the new emulsion 
since I have heard another report that at even lower doses the old emulsion 
could be pushed to get more OD with little increase in fog level.  The new 
emulsion does not seem to have that ability.
         Bob